Detailed Review:

I reviewed the Texas Instruments DLP NIR Scan Evaluator Module (EVM)

Introduction

 

 

This Eval module is meant to get you up and running and familiar with the capabilities of this nifty device.

image

Near IR (NIR) refers to the 700nm to 2500 nm range of the light spectrum. Outside of our visible spectrum, this range of light is used for spectroscopy in a variety of industries. Spectroscopy can allow identification of different materials based on the unique "fingerprint" they have when illuminated by a light source and having optical properties measured such as transmittance, reflectance, or absorbance. The degree of transmission or absorbance at each wavelength can create a fairly unique "fingerprint" to tell different materials and compounds apart.

 


NIR Spectroscopy has applications in several industries, including medical, pharmaceutical, agriculture, food production and QA, non-destructive evaluation (NDE), chemical production, chemistry, and atmospheric science. I dialed into the TI Sales call where they made the case for this device, and provided great explanations on its competitive advantages.As with all spectrometer, a sample is illuminated by a broadband (multi spectral) light source. The light reflecting off the material is directed, focused, split by a diffraction grating, and then received on some sort of sensing surface. Convention spectrometers frequently employ a linear array to sense each wavelength at a different point on the sensing surface.

image

 

 

Hardware

 


This DLP NIRScan spectrometer can sense in the typical 700-2500 nm wavelength range, but while built on this EVM, it can only sense in the 900-1700nm range. The manual talks about many ways to interface this hardware differently. I'm a little foggy in this area, but it appears you can interface it with your own logic controllers or microcontrollers and make complete custom use of this device. The manual and files talk about some RTOS and other softwares that enable you to get started. Reading deeper in the manual, the EVM board has a microcontroller running a Texas Instruments OS called Tiva, which runs on RTOS. The EVM board also contains micro SD card for saving data, Bluetooth adapter for interfacing via apps in iOS or Android devices, and a USB port for computer based interfacing.
The actual DLP detector board contains 6 main components. The moneymaker is a 1mm non-cooled Hamamatsu G12180 Indium Gallium Arsenide InGaAs detector. A TMP006 Thermopile will measure temperature of the detector and so that temperature compensation can be performed. Signal from the InGaAs sensor is amplifed by an OPA2376 low noise amplifier, and then converted into a 24 bit value by a ADS1255 ADC. Two other precision ICs provide precision voltage buffering and supply to the transimpedance amplifier. These are a REF5025 and OPA350. In short summary, this EVM module consists of 4 boards. The EVM microcontroller board (the one connect to via bluetooth or USB), DLP controller board (logic board that controls detector and contains lamp driver), detector board (board that does sensing and ADC), and the DMD board (contains DLP NIR mirror).  From a brief look at the manual, it looks like you'd want to interface the J500 port directly on the DLP controller board if you wanted to replace/bypass the EVM microcontroller and take full control of the sensing hardware. More conventionally, you could send commands to the TI EVM microcontroller board via UART4 on the J3 expansion connector or the J4 JTAG ARM Cortex 10-pin. The EVM contains 3 hardware buttons that are frankly a pain to access. All buttons are at the edge of the PCB but basically require you to cradle this little device and jam your fingernails in. Trying to get in there with a plastic tool probably still leaves you at a risk of jabbing the device. Luckily you really only need the scan/bluetooth button when you want to get up and running on Bluetooth.

image

 

 

Setup (Software, connection, app)

 


Installing the software is easy enough. Go to http://www.ti.com/tool/dlpnirnanoevm and look around for a good list of manuals, install files, etc. The page could really use a clean-up, as many older files are mixed in with new files, and clutter up the web page. Overall, it easy to gain access to a lot of material, including the hefty 100 PDF user's guide. But the website is a bit overwhelming if you want to just plug in the device, install the GUI, and start scanning items.

image

 

In not reading this page properly the first time, I installed an out of date GUI.

image

 

The Update GUI has a lot more bells and whistles when you install the correct one!

image

 

Installing the software goes well enough, except for some out of date Firmware files.

image

 

If you really care, you can go back into the website, download more install files, go into the extract folders, and then point to the ISO and BIN files to update your firmware. This part is not very user friendly and poorly thought out.

 

When you open the software, you find a pretty well laid out application. There is a quick start page that arrives in the package for those eager engineers (me) who hate manuals and just want to play with the tech. It does a good job of getting you rolling and doing a first scan.

 

The manual and webpage mentions that there is an iOS app available for reading from this device. There is also an Android App available, but it must be quite new because it is not mentioned anywhere at all. You can find it here: https://play.google.com/store/apps/details?id=com.kstechnologies.NanoScan&hl=en. To setup with the app: Simply open the app on phone, turn on bluetooth, and then hold the Bluetooth/Scan button on the device for 3 seconds. When you see the blue light flashing, you can press the little icon in upper right that looks like an electrical cord connecting. Connection should happen pretty quickly. When you load, you notice they have pre loaded some different NIR signatures, to show you what the different spectra can look like, and to give you some ideas on things to go scan.

image

 

 

 

Let's get to actually scanning some goodies! The quick start recommends you just use the stock scanning parameters. All you need to configure is the save location and filename. Click Scan, and the device takes care of the rest. As mentioned above, the device will bath the mystery object in broad spectrum light (probably like an incandescent type light source), and the measures the light reflected by the object in each wavelength. The intensity of this received light is registered. From the intensity of this received light the device determines the light reflected by the object, and the corresponding light transmission through the sample can be determined.

 

My first scan was of an IR remote. It's a cheap little one with an IR LED, and I honestly didn't know what the beam profile would look like. I see the expected peak in the 900 nm range, but see several peaks, which I didn't really expect. At first I wondered if the light beam is coming out "choppy" due to the diffraction grating of the NIR DLP device. But I expect this is the true output, and that this is perhaps the noisy IR output of the very cheap IR remote.

image

 

I hooked up some other IR LEDs in a basic regulated circuit to verify the emission spectra. Below is a TSAL6200 IR LED. It should have a peak wavelength at 940nm.

image

 

I do the scan and get the expected intensity peak at 940 nm. This device is working! The software shows a cool display of the detector senses light reflected or absorbed, and also the light intensity received at each wavelength. What isn't working is this GUI software. The stock GUI software outputs a CSV of only the absorbance values, which are not interesting at all for an IR LED.

 

image

 

 

Testing some other materials on hand, we can see some interesting patterns.

Testing

I set up a basic testing setup in my kitchen and basically tested all the organic components of a charcuterie board, and different alcohols, for good measure. Because the device runs off a standard 5V USB, you can easily power it with a battery pack, and then control scans on your phone via Bluetooth connection. The app allows me to configure saving to the DLP micro SD card port or the phone. Very handy out of the box! Below I am scanning salt water.

 

image

 

This was definitely a very fun device to really dive into. After my software / firmware concerns, I was really able to get up and running, and start testing different fruits, vegetables, cheeses, and wines in my kitchen. This is definitely science. Delicious science.

 

image

 

Plant health

 

Looking at wet and dried Basil leaves, we can clearly see a difference. It's likely that the water content (and possibly other plant health factors) can be differentiated by the different NIR signature.

image

image

 

 

Not being much of a spectroscopy expert, I was a little lost on the difference between comparing intensity and absorbance. From looking at the manual, I think intensity might essentially be the raw power levels recording with the DLP. I think the absorbance (and correspondingly the reflectance) are a calculated, and better normalized value. With that in mind let's use absorbance values moving forward.

 

My wet basil has a peak absorbance at 1400 nm, while the dry basil has overall much lower absorbance values (look at y axis), and peaks are poorly pronounced, but visible at 1400nm and 900 nm. We can see a clear difference.

 

image

image

 

Comparing different types of veggies

 

Below is an orange. Honestly is visually hard to tell all of the produce apart just by looking at these spectra, but with a mathematical analysis I'm sure you could develop a decent finger printing method. In the image below I can see that there is a small local max just past 1580 nm, probably around 1650nm. I also see a deeper trough around 1085nm, or higher peaks around 1150nm.

image

 

 

Let's scan some fresh food and their packaged counterparts

 

Here is a fresh apricot. Right below it is apricot jam. Note that the signal appears noisier for the jam, but this is likely due to me scanning it through the glass jar (not wanting to contaminate my EVM). We can see that the 1400nm peak is relatively reduced, and there are higher readings across 1100-15nm. Part of this may be due to the glass (borosilicate glass should have low absorbance up past 2000nm), but a this should mainly be caused additives to the jam, such as sugar, pectin, and citric acid preservative.

 

image

image

 

 

 

Liquids

Let's review a couple other materials, and look at some liquids. I tried to take a NIR scan of regular water in a glass as a good baseline, but again ran into issues with my sample jar. I guess a well designed flat sample container would help mitigate these issues.

 

 

 

image

Looking at salt water, we see following absorbance.  Slight peak around 1400-1500nm, but generally quite flat. Noise is probably an issue.

image

 

Possibly because the bottle was quite clear, we got a great reading from our scotch sample.

Absorption is highest at 1085nm, and has a low point around 1150 nm.

image

 

I took a sample of black coffee, and got a different spectra. Their app has a spectra for coffee available, which ended up being very different from mine. This may be due to my extremely small sample size, as I was just emptying the remnants of my coffee maker.

image

 

 

image

 

A couple other items with markedly different absorption spectra.

Honey shows very different absorbance, with peaks at 950nm and 1200nm.

image

 

Flour shows strong absorbance only around 1400nm.

 

 

image

Crystals

 

Let's look at some sugar samples

White sugar has a sharp peak at 1700nm ish and a noticeable peak around 1450nm.

image

 

For Brown sugar we notice a very different peak profile, with a broad peak that spans 1450 to 1600nm.

 

image

 

Why all the peaks around 1400 nm?

 

In general I am seeing that most materials have a noticeable absorbance around 1400nm. Looking at spectral absorbance of light in our atmosphere I notice that almost all light in 1400nm range is blocked by water moisture in the air. This seems to be our explanation. The 1400 nm peak is attributed to water. In the dry basil, we see the remaining leftover water as a small peak but it is a relatively small peak compared to the baseline signal.

 

image

 

Conclusion

 

In summary, this was a great piece of hardware to get exposed to the hands on science of taking NIR spectroscopy measurements. This is a well designed piece of hardware, and the user is able to get the system up and running quickly, and take their first spectra readings within the hour. Most of the confusion comes from their online website, which confusingly serves up multiple versions of files, and there isn't even a newest version that contains all software and firmware together with the GUI. This one roadblock is enough to waste time of the purchasing engineer, who may be left questioning the device. But once you get the device up and running, this is a slick little product. Future updates to this product should include updates to smooth out the install process, the GUI layout itself, and the CSV export in windows, which at the point of writing still only outputs absorbance values only. The EVM pack would be even more useful if the sensor probes were included in the EVM kit. These probes make liquid probing a lot easier, removing the need for a special glass test vials or slides for testing.

 

A couple lessons learned from myself: I need a better method of investigating wet samples. You don't want to damage the module, so you need a safe and reliable way of getting a wet sample right up against sensing window without contaminating or damaging the unit. Most flasks and test tubes are not ideal, since they are curved and I noticed that this really messed up readings. The ideal scenario is to get an external probe so that you can in close without risking damage to the unit (just like they would be doing in industry). Alternatively for small batch tests of wet materials, some sort of flat glass container or a test slide could work. I guess you could even put the wet sample in plastic wrap, but I wouldn't trust that. Below is an example of my folly in scanning my red wine in a curved glass. Even if I get in close to the sample, there is so much light scattering occurring, which seems to generally lead to a very noisy signal output, and depression of the real peaks. My only guess is that this geometry is really affecting the path length of the beams, and that different amounts of transmission in the glass medium are really distorting results.

 

image

 

Summary

 

Strengths

  • Unit works well once connected
  • Connects easily to Bluetooth
  • Quality Android App

 

Improvement Areas

  • Clean up website with quick links to the latest install files
  • Streamline install files, include all new firmware & software together
  • Update GUI (CSV export of all absorbance, reflectance, intensity values)
  • Include best practices for sample positioning (to improve user success with initial readings)
  • Allow normalized absorbance values to be output (making for faster comparison of different scans)
  • Identify major peaks in desktop GUI and mobile app, allowing users to quickly compare different scans
Anonymous
Parents
  • Nice review.

    As for the choppy spectra from IR remote - I also noticed this, but in my opinion that does not depend on the quality of IR LED. I noticed that this device does not work well with pulsed radiation. When I tried to measure my IR remote, peak was moving from 940nm to 1600nm. It was very inconsistent. The remote worked at 37kHz if I remember correctly. I will perform a better analysis on this subject, perform scan at different frequencies.

  • Thought about the pulsing problem a bit more. I've looked at the waveform of my specific IR remote, and I can see that it sends several square pulses and then has a large dead period. Couldn't find a picture off-hand but it's similar to the image below.

     

    So depending on when the spectrometer takes the reading, it may be sampling at a period where the LED is actually off. I guess the easiest fix is to take a series of readings, and use some software to combine them and fill in the blanks.

     

    image

     

    Here is image of my "chopped result" from my IR remote. Seems like the high points are regularly spaced. Also, the low points are not complete zeros. So I am still quite unsure about all of this image

    image

Comment
  • Thought about the pulsing problem a bit more. I've looked at the waveform of my specific IR remote, and I can see that it sends several square pulses and then has a large dead period. Couldn't find a picture off-hand but it's similar to the image below.

     

    So depending on when the spectrometer takes the reading, it may be sampling at a period where the LED is actually off. I guess the easiest fix is to take a series of readings, and use some software to combine them and fill in the blanks.

     

    image

     

    Here is image of my "chopped result" from my IR remote. Seems like the high points are regularly spaced. Also, the low points are not complete zeros. So I am still quite unsure about all of this image

    image

Children
  • I would check on the various codes used in IR remotes.

    The serial code is standard across devices, so you should be able to resolve the output.

     

    I am still kind of puzzled as to why that would cause any spectral pattern unless they have added multiple doped materials in the substrate to cover a wider range of input filters for better spectral response.

     

    It could be worth contacting the manufacturer of the device to see if they will provide some information.

     

    DAB

  • Sorry about that, Jan.

    I copy/pasted from my google album and it seems to be an embedded link to the photo.

     

    Should be fixed now!

  • Jordan Lui wrote:

     

    ...

     

    Here is image of my "chopped result" from my IR remote. Seems like the high points are regularly spaced. Also, the low points are not complete zeros. So I am still quite unsure about all of this

    image

    That image isn't visible for me, Jordan