Member's Forum Keysight 33250A Arbitrary Waveform Generator Fix
  • Replies 15 replies
  • Subscribers 627 subscribers
  • Views 3630 views
  • 33250A
  • keysight
  • arbitrary waveform generator
  • awg
charlieo21
charlieo21 over 3 years ago
I just want to share the score that I did with this arbitrary waveform generator that wasn't working properly.
It failed at amplitudes from 1Vpp to 10Vpp. Also, it failed the self-test, the seller told me that it just freezes when running the self-test.
I ran the self-test at night and next day it showed that failed the self-test with these two errors:
629: +20 dB amplifier path
630: Internal ADC overranged during one of the above tests
With this information, I decided to fix it, I found it quite easy to open it, a few screws and that's it.
It has a lot of passives in the bottom side and a lot of ICs in the top side, the power supply has its own board and apparently it wasn't design by Keysight.
I found the service manual and I started debugging it.
This table show what attenuators and/or amplifier is used in the different ranges.
The hybrid amplifier is used for the higher amplitude ranges. I was a little scared that the hybrid amplifier was damaged, and the AWG was done.
I started pocking around the inputs and output of the hybrid amplifier.
When I place the probe in the pin 6 (-INP) the output it showed the correct amplitude.
Basically, the pin connection wasn't making a good connection.
I pushed the pin just to be sure that it's doing a good connection with the hybrid.
Now it's working and it passed the self-test.
    DAB over 3 years ago
    Well done.
    I love to see simple fixes work.
  • Gough Lui
    Gough Lui over 3 years ago in reply to charlieo21
    Fair enough, although to me it is still modern as it has an LCD :)
  • Gough Lui
    Hmm. I'd say heating and cooling cycle may well change the calibration in some measurable way ... I'd be a bit nervous about doing it if it wasn't entirely necessary.
    - Gough
  • dang74
    dang74 over 3 years ago
    Nice work.  It's so rewarding when you get to apply your knowledge to successfully fix something.
  • charlieo21
    charlieo21 over 3 years ago in reply to Gough Lui
    Well it's not exactly modern, I think the first models were build in the late 90s. This model has been discontinued in 2016. Definitely it was nice to find that the service guide with schematics and theory of operation was available online.
  • charlieo21
    charlieo21 over 3 years ago in reply to dougw
    I thought on putting some solder in the pins but probably it would change the impedance and distort the signals.
  • Gough Lui
    It's been rare to see any modern gear with schematics ... glad you were able to get them and get it back to life!
    dougw over 3 years ago
    It is awesome when your confidence pays off.  Does the flaky connection need a more permanent fix?
  • colporteur
    colporteur over 3 years ago
    Always nice to solve a dog fault (i.e. persistent) with a small dodo bag (i.e simple inexpensive repair).
  • vladimir_popoff
    vladimir_popoff over 3 years ago
    He's very important to my lab.