Ask an Expert Forum Clarification on AD7192 System Calibration Input Span vs Observed Full-Scale Behavior
  • +1 person also asked this people also asked this
  • Subscribers 338 subscribers
  • Views 189 views
  • EVAL-AD7192ASDZ
  • ad7192
See a helpful answer?
Jay_Instrument
Jay_Instrument 5 months ago
We are currently evaluating the AD7192 ADC as a replacement for the obsolete CS5532 in our load cell application.
With the CS5532, after system calibration, we were able to achieve full-scale output at an input of 10 mV. While testing the AD7192, we are observing behavior that does not fully align with the system calibration input span specifications given in the datasheet.
Our configuration is as follows:
  • Reference voltage (Vref): 5 V
  • Gain: 128
  • Calculated full-scale input range (FS): 39 mV
As per the datasheet, the allowable input span for system calibration is:
  • Minimum input: 0.8 × FS ≈ 31 mV
  • Maximum input: 2.1 × FS ≈ 81 mV
However, our load cell has a maximum output of 10 mV, which is significantly below the specified minimum calibration span.
Observed behavior:
  • We tested inputs at 10 mV, 15 mV, and 30 mV.
  • We begin to observe full-scale counts starting at approximately 15 mV.
  • This is notably lower than the expected ~31 mV minimum based on the datasheet.
Questions:
  1. What could explain achieving full-scale counts at ~15 mV, which is below the specified minimum system calibration span?
  2. Is this behavior valid and reliable for production use, or could it lead to accuracy or stability issues?
  3. Given our load cell output (max 10 mV), would you recommend proceeding with AD7192, or should we consider alternative configurations or devices?
Additionally, could you please connect us with an Analog Devices application engineer to further investigate and discuss this issue in detail?
We would appreciate your guidance in understanding whether this observed behavior is within acceptable operation limits and suitable for long-term use.
    michaelkellett 5 months ago +3 verified
    You need to describe your set up in more detail. Check out the Data Output Coding on page 32 of the data sheet. Are you working in uni-polar or bipolar mode. Another possibility is that you are nor…
    +1 michaelkellett 5 months ago
    Another possibility is that you are nor reading the device output correctly but are shifting the data by 1 bit at some point.
    You should be aware that E14 is a forum with a wide range of members and while you may get direct support from Avnet/Farnell/Newark you may not.
    If you want support from AD you should contact them directly.
    I don't work for any of the above and my suggestions are based on general experience of load cells and ADCs and looking at the device data sheet.