Test & Tools
Test & Tools
Forum If you need help on Test and Measurements, please post your question here!
  • Replies 48 replies
  • Subscribers 389 subscribers
  • Views 5829 views
  • test
  • 测试
  • application
  • and
  • measurement
  • oscilloscopes
  • measurements
  • fluke
  • 示波器
This thread has been closed to new questions.
However, we welcome you to Post Your Question about Test & Measurement in the element14 Community Test & Measurement group. You'll find many fellow members and experts who have just the answer you're looking to find!
Thank You, Your Friends at element14 Community
Jim Newell
Jim has worked for Fluke for a little over 31 years working in Manufacturing production and Quality assurance for the first 15 years. In the last 15 years he has focused on technical support as a Application Engineer for all of the Electrical Industrial products such as Handheld Oscilloscopes, Digital Mulitmeters and Power Quality Analyzers.  He enjoys the challenge of helping people and the famous saying the only dumb question is the one you didn’t ask are often joked about among his customers.
    e14 Contributor over 15 years ago in reply to mouch
    Hi Dirk,
    You have a few options, each with pros & cons:
    1) Use a differential probe to measure across each phase.
    2) Use a scope with isolated channels.  Several handheld models offer that capability, plus the Tektronix TPS2000 series instruments.
    3) Measure each phase relative to ground and use maths function on the scope, or offline ona PC) to derive the phase measurements.
    In any case DON'T cut your scope's ground lead and float the instrument - it's dangerous and violates numerous safety standards.
    Tektronix have an application note on making floating measurements.  You can download it from: http://www.tek.com/Measurement/App_Notes/3M_17773/eng/3MW_17773_0.pdf
    Cheers!
    hi Jim,
    I have to implement the 10MHz reference signal project for our measurement devices.
    I would like to know if Tektronix does have GPS disciplined Rubidium oscillator devices  + antenna + surge protection + redundancy option
    thanks!
  • dsohal
    dsohal over 15 years ago in reply to e14 Contributor
    I am sorry, Tektronix do not have GPS disciplined oscillator devices and I have checked in Farnell range as well with other our supplier.
    Thank you for enquiry.
    Best regards,
    Farnell technical
    e14 Contributor over 15 years ago in reply to dsohal
    Hi Devinder,
    Tektronix offer a line of frequency counter/timers - the FCA3000FCA3000 & MCA3000 - but not the Rubidium oscillator devices you are seeking.
    Try contacting Quartzlock Ltd.  They may have a product that meets your needs. http://quartzlock.com/
    Trevor
    Trevor J. Smith|  Technical Marketing Manager  |  Phone: +44 (0)1344 392225 |  Mobile: +44 (0)7836) 516853  |  www.tektronix.com<http://www.tektronix.com/>
    For Tektronix legal entity information and registration details go to: www.tektronix.com/entities<http://www.tektronix.com/entities
    I have Fluke 5102B calibrator,which is working fine with its D.C ranges but it does not works with its A.C. ranges.The o/p is exactly 2.14 times higher than the programmed value.Readings are as follows:
    Programmed values at 1kHz           o/p @fluke 8060A DMM
    100mV                                              0.2128V,967.7Hz
    1V                                                     2.135  V,967.7Hz
    10V                                                   21.40  V,967.6Hz
    100V                                                 213.9  V,967.7Hz
    400V                                                 867.2  V,967.5Hz
    100microAmps                                  0.2136microAmp,967.7Hz
    1mA                                                   2.133mA
    10mA                                                 21.36mA
    100mA                                               213.1mA
    500mA                                              1070.0mA
    It's Ohm ranges are showing stable & correct for 10,100,1k,10k Ohms,but for 100k,1M,10M does not stable.It varies around its expected values.
    For 100kOhms o/p varies between 99.79 to 100.22kOhms same for 1M &10M
    Manufacturing date of the calibrator is around 1987.
    Can you guide me which component exactly I have to replace.
    Thanks with regards
    KIRAN
    xepa08@gmail.com
  • dsohal
    Hi Kiran,
                I am sorry,I was away, unable to reply you and regarding problem with Fluke calibrator 5102 B, I am sorry,you have to contact Fluke and their contact Tel 01603256690, because we do not deal with any repair or calibration.
    Thank you for your enquiry,
    Devinder Sohal,
    Farnell technical,
    Hi Jim,
    I am trying to calculate a Mean Time Between Failures (MTBF) for a system using MIL HDBK 217F, unfortunately I still have a few components left which I can't seem to find the MTBF for, nor do I know how I should categorise them to use the MIL HDBK to get an MTBF for them.  The components I am struggling with are:
    1. Octal Buffer, Line Driver.
    2. Amp Isolation
    3. Buffer / Line Driver
    4. Temp Sensor
    5. Power Distribution Switch
    6. CPCI Bus Hot Swap Controller
    I have included links for the components below.  If you could tell me what section these fall under in the MIL HDBK or anything else that would help me predict the MTBF you would be doing me a huge favour.
    http://au.element14.com/texas-instruments/sn74hct541dw/ic-octal-buffer-line-driver-smd/dp/9592059?Ntt=959-2059
    http://au.element14.com/texas-instruments/iso124u/ic-amp-isolation-smd-soic28-124/dp/1212426?Ntt=121-2426
    http://au.element14.com/texas-instruments/sn74ahct1g125dbvr/ic-74ahct-single-gate-smd/dp/1105927?Ntt=1105927
    http://au.element14.com/texas-instruments/tmp275aidgkt/ic-temp-sensor-5-5v-smd-msop8/dp/1470503?Ntt=1470503
    http://au.element14.com/texas-instruments/tps2024dg4/power-distribution-switch-soic8/dp/1235040?Ntt=1235040
    http://www.linear.com/product/LTC4244
    Thank for your time,
    Peter
    HI Jim
    Would you have any recommendations on a replacement for the Fluke 8920A true RMS RF voltmeter ?
    My requirements are true RMS millivolt measurements centered around the 400kHz region.
    The nearest I have found is the Rohde&Schwarz URE3 but this is costly, and the Fluke 85RF probe adaptor is not sensitive enough.
    Any ideas suggestions ( that don't involve buying second hand 8920A's ) welcome.
    Regards David Mordey.
    Jim,
    Can you tell me how to capture the milliseconds time to trip RCD Circuit Breakers? I believe you would need a trigger point to start the timer and store the time value once the shunt trip has opened the CB. Can this be done without a Power Oscilliscope?
    I have a Fluke 9100 Calibrator but it is limited for RCD tests.
    Basil
    Jim,
    I would like some help in trying to find the right firm to develop a light meter that currently is not manufactured by any vendor. We have some very specific requirements for a meter used in a niche business.
    ~Cody