element14 Community
element14 Community
    Register Log In
  • Site
  • Search
  • Log In Register
  • Community Hub
    Community Hub
    • What's New on element14
    • Feedback and Support
    • Benefits of Membership
    • Personal Blogs
    • Members Area
    • Achievement Levels
  • Learn
    Learn
    • Ask an Expert
    • eBooks
    • element14 presents
    • Learning Center
    • Tech Spotlight
    • STEM Academy
    • Webinars, Training and Events
    • Learning Groups
  • Technologies
    Technologies
    • 3D Printing
    • FPGA
    • Industrial Automation
    • Internet of Things
    • Power & Energy
    • Sensors
    • Technology Groups
  • Challenges & Projects
    Challenges & Projects
    • Design Challenges
    • element14 presents Projects
    • Project14
    • Arduino Projects
    • Raspberry Pi Projects
    • Project Groups
  • Products & Partners
    Products & Partners
    • Arduino
    • Avnet & Tria Boards Community
    • Dev Tools
    • Manufacturers
    • Multicomp Pro
    • Product Groups
    • Raspberry Pi
    • RoadTests & Reviews
  • About Us
    About the element14 Community
  • Store
    Store
    • Visit Your Store
    • Choose another store...
      • Europe
      •  Austria (German)
      •  Belgium (Dutch, French)
      •  Bulgaria (Bulgarian)
      •  Czech Republic (Czech)
      •  Denmark (Danish)
      •  Estonia (Estonian)
      •  Finland (Finnish)
      •  France (French)
      •  Germany (German)
      •  Hungary (Hungarian)
      •  Ireland
      •  Israel
      •  Italy (Italian)
      •  Latvia (Latvian)
      •  
      •  Lithuania (Lithuanian)
      •  Netherlands (Dutch)
      •  Norway (Norwegian)
      •  Poland (Polish)
      •  Portugal (Portuguese)
      •  Romania (Romanian)
      •  Russia (Russian)
      •  Slovakia (Slovak)
      •  Slovenia (Slovenian)
      •  Spain (Spanish)
      •  Sweden (Swedish)
      •  Switzerland(German, French)
      •  Turkey (Turkish)
      •  United Kingdom
      • Asia Pacific
      •  Australia
      •  China
      •  Hong Kong
      •  India
      •  Japan
      •  Korea (Korean)
      •  Malaysia
      •  New Zealand
      •  Philippines
      •  Singapore
      •  Taiwan
      •  Thailand (Thai)
      •  Vietnam
      • Americas
      •  Brazil (Portuguese)
      •  Canada
      •  Mexico (Spanish)
      •  United States
      Can't find the country/region you're looking for? Visit our export site or find a local distributor.
  • Translate
  • Profile
  • Settings
Restoration & Repair
  • Challenges & Projects
  • Project14
  • Restoration & Repair
  • More
  • Cancel
Restoration & Repair
Blog PSU accuracy measurement and calibration
  • Blog
  • Forum
  • Documents
  • Events
  • Polls
  • Files
  • Members
  • Mentions
  • Sub-Groups
  • Tags
  • More
  • Cancel
  • New
Join Restoration & Repair to participate - click to join for free!
  • Share
  • More
  • Cancel
Group Actions
  • Group RSS
  • More
  • Cancel
Engagement
  • Author Author: neuromodulator
  • Date Created: 19 Apr 2020 7:08 PM Date Created
  • Views 7738 views
  • Likes 8 likes
  • Comments 43 comments
  • restorerepairch
Related
Recommended

PSU accuracy measurement and calibration

neuromodulator
neuromodulator
19 Apr 2020

  • Introduction
  • Pre-calibration accuracy
  • Calibration
  • Post-calibration accuracy
  • Source code
  • Conclusions

 

WARNING: DO NOT TRY TO REPLICATE THIS UNLESS YOU ARE WILLING TO TAKE THE RISK OF BREAKING YOUR PSU!

 

 

Introduction

 

Rigol recommends calibrating the Rigol DP832 power supply unit (PSU) (Figure 1) at least once a year, and since my PSU's calibration date was 5 years old, I thought it would be a good time to recalibrate it. Before calibrating the PSU, I decided to evaluate it so that I could check if it was still in spec.

 

image

Figure 1: Rigol DP832 [Source: Rigol].

 

 

While looking for the specs, I found some discrepancies between the annual accuracy shown in the datasheet (Table 1) and in the performance verification manual (Table 2):

 

image

Table 1: Annual accuracy according to the datasheet [Source: Rigol].

 

image

Table 2: Annual accuracy according to the performance verification manual [Source: Rigol].

 

 

So I arbitrarily chose to test the instrument against the following accuracy limits:

 

Item

Annual accuracy (25℃ ±5℃)

±(Output percentage + offset)

Voltage programming 0.05% + 10mV
Voltage readback 0.05% + 5mV
Current programming 0.2% + 5mA
Current readback 0.15%+ 5mA

 

 

My previous experience with Rigol's test equipment software and remote control was not good, they were unstable and not usable at all. I was surprised to find that now the TCP/IP link worked very well.

 

To control the PSU I used Python with PyVisa and the following simple PSU SCPI commands:

  • *RST

         Restores the PSU to the factory settings and clears the error queue.

 

  • :OUTP

        Enables or disables the output of the specified channel.

 

  • :APPL

        Sets the voltage and current for the specified channel.

 

  • :MEAS:ALL?

        Queries the voltage, current and power measured on the output terminal of the specified channel.

 

 

The PSU was connected to a Keithley DMM6500 which was controlled through simple TSP commands (which I've already shown in my Keithley Bench Digital Multimeter - Review). Two type of connections have to be made for each channel, to measure voltage (Figure 2) and current (Figure 3).

 

image

Figure 2: Voltage measurement [Source: Rigol].

 

 

image

Figure 3: Current measurement [Source: Rigol].

 

 

Each channel contains 4 calibration items:

  • Programmed voltage (V-DAC)
  • Programmed current (I-DAC)
  • Readback voltage (V-ADC)
  • Readback current (I-ADC)

 

To measure the accuracy of the instrument, 12 calibration items need to be evaluated.

 

 

Pre-calibration accuracy

 

Accuracy was measured by performing full voltage and current sweeps at each channel with a 1 s delay at each sweep step and measurement. 0.1 V steps were used in the voltage sweep, and 0.01 A steps in the current sweep. The following table shows the obtained results:

 

Pre-cal      V-DAC      V-ADC       I-DAC      I-ADC

------------------------------------------------------

Channel 1:   ±12.8 mV    ±8.1 mV    ±0.8 mA    ±1.5 mA

Channel 2:   ±12.4 mV   ±10.4 mV    ±2.2 mA    ±2.0 mA

Channel 3:    ±4.6 mV    ±2.5 mV    ±4.1 mA    ±3.6 mA

 

 

The following 2 figures show the programmed and readback voltage and current errors:

imageimage

 

Figure 4: Programmed voltage and readback voltage error plots.

imageimage

Figure 5: Programmed voltage and readback current error plots.

 

 

Everything looked fine, with the exception of the channel 2 readback voltage, which at 0 V gets slightly out off spec. The reason that it gets out of spec is that when the PSU is programmed to output 0 V, it actually outputs ~-8 mV, which is then read back as a clamped 0 V. A similar situation occurs with channel 3 in the readback current plot, but it does not get out of spec there.

 

 

Calibration

 

Rigol provides a calibration manual that shows how to manually calibrate the PSU. To calibrate it, the 4 calibration items (V-DAC, V-ADC, I-DAC, I-ADC) need to be evaluated at different "calibration points" (Figure 6). When a "cal point" is selected the PSU generates an arbitrary voltage/current output that depends on the cal point ID, this output has to be measured with a DMM and manually entered in the table. The procedure needs to be repeated for every "cal point" of the table until it's complete. Since the PSU has 3 channels and there are 4 calibration items, 12 tables need to be filled to fully calibrate the PSU. There is a total of 180 values that have to be entered, which makes the process slow,and error prone.

 

image

Figure 6: Calibration table [Source: Rigol].

 

 

Luckily there is a better approach through the usage of SCPI commands. Members of the eevblog reverse engineered the PSU calibration SCPI commmands and published an SCPI calibration user guide. Although a good starting point, this guide has some minor errors and misses some important details, such as that the number of calibration points and values can be set arbitrarily. Lets take a look at the commands that I used on my calibrator program:

  • :CAL:START

        Begins the calibration of the specified channel.

 

  • :CAL:END

        Ends the calibration and saves the specified date.

 

  • :CAL:CLEAR

        Clears the calibration table.

 

  • :CAL:SET

       Sets the PSU output to the specified voltage/current and enters a calibration point output value in the table.

 

  • :CAL:MEAS

       Sets the DMM measured voltage/current value to the specified calibration point.

 

 

The number of calibration points, and their output values were set by trial and error. Some calibration point values increased the error at the low end of the voltage/current values in a non-trivial way. After some testing, I finally decided to used a total of 386 calibration points to perform a full calibration.

 

Here is a text output sample of an ongoing calibration:

 

Connected to PSU at 'TCPIP0::192.168.0.35::inst0::INSTR'

Connected to DMM at 'TCPIP0::192.168.0.30::inst0::INSTR'

 

press Enter to begin calibration...

Calibrating Ch1 V-DAC

[1/65] V-DAC: 0.250, V-DMM: -0.05259

[2/65] V-DAC: 0.500, V-DMM: 0.19608

[3/65] V-DAC: 1.000, V-DMM: 0.69401

[4/65] V-DAC: 1.500, V-DMM: 1.19042

[5/65] V-DAC: 2.000, V-DMM: 1.68867

[6/65] V-DAC: 2.500, V-DMM: 2.18614

[7/65] V-DAC: 3.000, V-DMM: 2.68230

[8/65] V-DAC: 3.500, V-DMM: 3.17921

[9/65] V-DAC: 4.000, V-DMM: 3.67903

[10/65] V-DAC: 4.500, V-DMM: 4.17686

[11/65] V-DAC: 5.000, V-DMM: 4.67444

[12/65] V-DAC: 5.500, V-DMM: 5.17170

[13/65] V-DAC: 6.000, V-DMM: 5.66934

[14/65] V-DAC: 6.500, V-DMM: 6.16577

[15/65] V-DAC: 7.000, V-DMM: 6.66192

[16/65] V-DAC: 7.500, V-DMM: 7.15350

[17/65] V-DAC: 8.000, V-DMM: 7.64920

[18/65] V-DAC: 8.500, V-DMM: 8.14618

[19/65] V-DAC: 9.000, V-DMM: 8.64358

 

 

Post-calibration accuracy

 

After performing a full calibration its time to compare the pre-calibration with the post-calibration results:

 

Pre-cal       V-DAC      V-ADC       I-DAC      I-ADC

------------------------------------------------------

Channel 1:   ±12.8 mV    ±8.1 mV    ±0.8 mA    ±1.5 mA

Channel 2:   ±12.4 mV   ±10.4 mV    ±2.2 mA    ±2.0 mA

Channel 3:    ±4.6 mV    ±2.5 mV    ±4.1 mA    ±3.6 mA

Post-cal      V-DAC      V-ADC      I-DAC      I-ADC

------------------------------------------------------

Channel 1:    ±2.3 mV    ±0.5 mV    ±1.8 mA    ±1.0 mA

Channel 2:    ±4.2 mV    ±0.7 mV    ±1.4 mA    ±0.4 mA

Channel 3:    ±0.8 mV    ±0.4 mV    ±0.7 mA    ±0.4 mA

imageimage

Figure 7: Pre- and post-calibration programmed voltage error plots.

imageimage

Figure 8: Pre- and post-calibration readback voltage error plots.

imageimage

Fullscreen 0068.contentimage_134938.html Download
<html><head><title>Jive SBS</title></head>
<body><font face="arial,helvetica,sans-serif">
<b>Error</b><br><font size="-1">
An general error occurred while processing your request.
</font></font></body></html>

Figure 9: Pre- and post-calibration programmed current error plots.

imageimage

Figure 10: Pre- and post-calibration readback current error plots.

 

 

The post-calibration accuracy increased in all cases except one: the programmed current for channel 1, which I suspect was caused by temperature variations. I found the programmmd current calibration/measurement to be quite tricky because heating of the PSU affects the current output. And since PSU temperature varies quite a bit during normal operation, how close the output is to the programmed values varies during the operation of the instrument. For example if all of a sudden a lot of current is sourced, it will take probably a few minutes for the PSU output to stabilize.

 

 

Source code

 

All the code was uploaded to GitHub. The code structure is as follows:

 

Psu.py

Defines the class Psu that represents the DP832.

 

Dmm.py

Defines the class Dmm that represents the Keithley DMM6500. This is the class that should be modified if one wants to implement a different DMM.

 

Calibrate.py

Contains the function that performs the calibration.

 

Measure.py

Contains the function that performs the measurements.

 

ClearCalibration.py

Clears all the calibration tables (useful when tables get corrupted).

 

CalibrateSingle.py

Calibrates a single channel current or source programming and readback.

 

CalibrateAll.py

Calibrates the complete instrument.

 

MeasureSingle.py

Measures the accuracy of a single channel current or voltage source programming and readback, and then results on the screen.

 

MeasureAll.py

Measures the accuracy of the complete instrument, and saves the results into a file.

 

PlotAccuracy.py

Reads the "MeasureAll" output file and generates plots.

 

 

Conclusions

 

Instrument calibration can be expensive and require sending instruments to calibration labs. Here I showed an automated alternative that requires very little effort and produces acceptable results. I also showed a way to measure and evaluate the 4 calibration items to help decide when to recalibrate the PSU. All the source code will be made available on GitHub as soon as I fix some minor issues and test the programs in more depth.

  • Sign in to reply

Top Comments

  • Jan Cumps
    Jan Cumps over 6 years ago +2
    I have this PSU too. The DP832A version. Never calibrated, hesitant to do that myself. Reminds me that my Keithley DMM is due for calibration. That one I want to bring to Welectron in Germany - always…
  • Jan Cumps
    Jan Cumps over 6 years ago in reply to neuromodulator +2
    neuromodulator wrote: ... I'll soon upload the files, maybe you could give the measurement program a try to evaluate the accuracy of the PSU. You wouldn't have to modify a single line of code to get it…
  • ralphjy
    ralphjy over 6 years ago +2
    Might just be me, but I don't see any of the figures in your post - only the captions.
  • Jan Cumps
    Jan Cumps over 6 years ago in reply to neuromodulator

                  V-DAC      V-ADC      I-DAC      I-ADC
    ------------------------------------------------------
    Channel 1:    ±6.0 mV    ±2.6 mV    ±2.0 mA    ±0.8 mA
    Channel 2:    ±6.0 mV    ±3.3 mV    ±3.2 mA    ±2.9 mA
    Channel 3:    ±1.9 mV    ±1.2 mV    ±2.7 mA    ±0.9 mA

    • Cancel
    • Vote Up +1 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • neuromodulator
    neuromodulator over 6 years ago in reply to Jan Cumps

    Your curves look better than mine before calibration. Could you post the table? (its printed in text mode to the console)

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • neuromodulator
    neuromodulator over 6 years ago in reply to Jan Cumps

    Yeah that's pre-warming, but I'm talking about load-dependent heating. Try fast current sweeps and you will get different results

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • Jan Cumps
    Jan Cumps over 6 years ago in reply to neuromodulator

    New measurements with the original calibration table:

     

    imageimageimageimage

     

     

                  V-DAC      V-ADC      I-DAC      I-ADC
    ------------------------------------------------------
    Channel 1:    ±6.0 mV    ±2.6 mV    ±2.0 mA    ±0.8 mA
    Channel 2:    ±6.0 mV    ±3.3 mV    ±3.2 mA    ±2.9 mA
    Channel 3:    ±1.9 mV    ±1.2 mV    ±2.7 mA    ±0.9 mA

    • Cancel
    • Vote Up +1 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • Jan Cumps
    Jan Cumps over 6 years ago in reply to neuromodulator

    I've been reading Rigol's manual. They advise to warm up the supply for at least 30 minutes before starting. (mine has been on for hours now, same for the DMM).

    No advice though on what to do when it's warming up during the current calibration...

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
>
element14 Community

element14 is the first online community specifically for engineers. Connect with your peers and get expert answers to your questions.

  • Members
  • Learn
  • Technologies
  • Challenges & Projects
  • Products
  • Store
  • About Us
  • Feedback & Support
  • FAQs
  • Terms of Use
  • Privacy Policy
  • Legal and Copyright Notices
  • Sitemap
  • Cookies

An Avnet Company © 2026 Premier Farnell Limited. All Rights Reserved.

Premier Farnell Ltd, registered in England and Wales (no 00876412), registered office: Farnell House, Forge Lane, Leeds LS12 2NE.

Follow element14

  • X
  • Facebook
  • linkedin
  • YouTube