element14 Community
element14 Community
    Register Log In
  • Site
  • Search
  • Log In Register
  • Community Hub
    Community Hub
    • What's New on element14
    • Feedback and Support
    • Benefits of Membership
    • Personal Blogs
    • Members Area
    • Achievement Levels
  • Learn
    Learn
    • Ask an Expert
    • eBooks
    • element14 presents
    • Learning Center
    • Tech Spotlight
    • STEM Academy
    • Webinars, Training and Events
    • Learning Groups
  • Technologies
    Technologies
    • 3D Printing
    • FPGA
    • Industrial Automation
    • Internet of Things
    • Power & Energy
    • Sensors
    • Technology Groups
  • Challenges & Projects
    Challenges & Projects
    • Design Challenges
    • element14 presents Projects
    • Project14
    • Arduino Projects
    • Raspberry Pi Projects
    • Project Groups
  • Products
    Products
    • Arduino
    • Avnet & Tria Boards Community
    • Dev Tools
    • Manufacturers
    • Multicomp Pro
    • Product Groups
    • Raspberry Pi
    • RoadTests & Reviews
  • About Us
  • Store
    Store
    • Visit Your Store
    • Choose another store...
      • Europe
      •  Austria (German)
      •  Belgium (Dutch, French)
      •  Bulgaria (Bulgarian)
      •  Czech Republic (Czech)
      •  Denmark (Danish)
      •  Estonia (Estonian)
      •  Finland (Finnish)
      •  France (French)
      •  Germany (German)
      •  Hungary (Hungarian)
      •  Ireland
      •  Israel
      •  Italy (Italian)
      •  Latvia (Latvian)
      •  
      •  Lithuania (Lithuanian)
      •  Netherlands (Dutch)
      •  Norway (Norwegian)
      •  Poland (Polish)
      •  Portugal (Portuguese)
      •  Romania (Romanian)
      •  Russia (Russian)
      •  Slovakia (Slovak)
      •  Slovenia (Slovenian)
      •  Spain (Spanish)
      •  Sweden (Swedish)
      •  Switzerland(German, French)
      •  Turkey (Turkish)
      •  United Kingdom
      • Asia Pacific
      •  Australia
      •  China
      •  Hong Kong
      •  India
      • Japan
      •  Korea (Korean)
      •  Malaysia
      •  New Zealand
      •  Philippines
      •  Singapore
      •  Taiwan
      •  Thailand (Thai)
      • Vietnam
      • Americas
      •  Brazil (Portuguese)
      •  Canada
      •  Mexico (Spanish)
      •  United States
      Can't find the country/region you're looking for? Visit our export site or find a local distributor.
  • Translate
  • Profile
  • Settings
RoadTests & Reviews
  • Products
  • More
RoadTests & Reviews
Review Blogs Tektronix 3 Series MDO - Part 2: Rotor Reflectometer Calibration
  • Blogs
  • RoadTest Forum
  • Documents
  • RoadTests
  • Reviews
  • Polls
  • Files
  • Members
  • Sub-Groups
  • More
  • Cancel
  • New
Join RoadTests & Reviews to participate - click to join for free!
  • Share
  • More
  • Cancel
  • Author Author: three-phase
  • Date Created: 20 Dec 2019 10:21 AM Date Created
  • Views 2090 views
  • Likes 9 likes
  • Comments 5 comments
Related
Recommended
  • rso test
  • tektronix
  • rotor reflectometer
  • 53220a
  • 3 series mdo

Tektronix 3 Series MDO - Part 2: Rotor Reflectometer Calibration

three-phase
three-phase
20 Dec 2019

    • Introduction
    • Calibration Checks
    • Measurement captures
    • Generator Test
      • Update 18th January 2020
    • Conclusions

 

Introduction

 

For this blog I am going to use the 3 Series MDO to check the performance of a rotor reflectometer. This instrument is utilised on generator rotors to determine the presence of interturn faults within the rotor winding.

 

To do this the reflectometer applies pulses to each end of the rotor winding in alternating manner and an oscilloscope compares the reflected pulses. If the reflected pulses match, then there are no faults. If the reflected pulses are different then a fault is present. The location of the fault can be determined by measuring the time to the reflected pulses and calculating the delay in terms of winding length.

 

The performance of the rotor reflectometer is generally checked with a Keysight 53220A53220A counter, but I will aim to collect the readings alongside this with the 3 Series MDO and compare the results obtained.

 

image

 

Calibration Checks

 

Checks are made on the reflectometer to ensure that the frequency of the pulses, the pulse widths, the pulse amplitudes and the interval between the input and output pulses are within expectations. I will also reflect on which methodology provides the most efficient way to carry out the checks.

 

The reflectometer has two outputs, the first is the single transit which is the pulse fired from one end of the winding to the other. The second is the double transit which is the reflected pulse seen at the end where the single transit pulse is fired from. As the counter is two channel, I connect the double transit output to channel one and the single transit to channel two. These are then paralleled to channels one and two of the oscilloscope.

 

The oscilloscope does not need any specific setup in terms of bandwidth or memory depth, the input impedance is set to 1 MOhm to avoid loading the reflectometer. The amplitude and timebase are adjusted to suit as the tests are progressed. As I am working with two channels, the measurements are changed dependent upon the tests due to the limitation of four measurements on screen at any one time.

 

image

 

Examples of signals captured and measurements made on oscilloscope

 

{gallery} Rotor Reflectometer Measurements

image

Maximum Pulse Frequency Measurement

image

Minimum Pulse Width Measurement

image

Maximum Pulse Width Measurement

image

Pulse Amplitude at 50 Ohm Load

image

Pulse Amplitude at 1000 Ohm Load

image

Interval Measurement between Single and Double Transit Pulses

 

The largest difference between using the oscilloscope and the counter is that the information for both channels is available on the oscilloscope at the same time, where as on the counter, there is a constant need to switch between the two channels to obtain the readings for the test settings. This makes the checks more convenient and a little quicker, there is also less risk of confusion between the channels and writing down the wrong reading. However, the counter does offer a higher level of accuracy than the oscilloscope.

 

image

 

The frequency and pulse width measurements between the two instruments were comparable. The largest deviation was around 5% for the lowest pulse width measurement, this seems out of characteristic with the other time based measurements, so I need to go back and double check that there wasn't a reading or setup error.

 

Amplitude measurements also showed good comparison between the counter and oscilloscope, the largest deviation being a little under 3% at the 500 Ohm load. Sometimes ripple is observed on the output waveforms and it may be this that has caused some of the error.

 

 

image

 

Resolution for the amplitude measurements was the same between the counter and the oscilloscope. Again, the checks were a little faster using the oscilloscope as there was no need to flip between channels to record the values.

 

The time interval between the input and output pulses proved to be more accurate on the oscilloscope when the timebase was set to expand the waveforms more across the screen. The initial measurement with a higher timebase provided a slightly lower reading in comparison to the counter.

 

image

 

Overall the checks went well. Although the oscilloscope is less accurate, the reflectometer is not in itself the most accurate of instruments and therefore the speed and convenience of the oscilloscope out weighs the accuracy, but slower operation, offered by the counter.

 

Measurement captures

 

The next stage after the checks was to utilise the test delay line to record some reflectometer measurements. These were for comparison against the Rigol DG1052E oscilloscope currently used to capture the traces. The test box utilised during the performance checks has the ability to simulate both inter-turn and earth faults, that will change the reflectometer output.

 

Initially the reflectometer is set up and the width of the double transit pulse recorded, this was easy to do with the cursor function of the oscilloscope. The second channel is then enabled and the time to the start of the single transit pulse measured. These two values are important as they can be used to calculate the distance to a fault should one be observed.

 

{gallery} RSO Measurements

image

Double transit pulse measurement

image

Single transit pulse measurement

 

Faults can then be placed and the screens captured. The first fault applied is a short between two adjacent turns of the winding. This results in a double deflection in the waveform.

 

image

 

This shows up quite clearly on the oscilloscope, here the advantage of such a large screen can clearly be seen. There is another significant improvement though, the screen capture has recorded both of the deflections. For some reason, the Rigol oscilloscope does not record both deflections, so I have to save one deflection as a reference waveform and then screen capture this alongside the channel waveform.

 

{gallery} Rigol Screen Captures

image

Rigol scope measurement of double transit pulse

image

Rigol scope capture of first deflection

image

Rigol scope capture of second deflection

image

Rigol scope display of both deflections using a reference waveform

 

This offers a huge advantage over the Rigol oscilloscope when carrying out reflectometer measurements with a rotating generator shaft. As the generator shaft slows down from 3000 rpm, the shape of the waveform changes slightly, using the Rigol, I have to repeatedly correct the reference waveform to get accurate readings, but with the 3 Series MDO this is not required and I can record the screen much more easily.

 

Lets simulate an earth fault and also record the pulse starting point.

 

image

 

Here one of the signals is dragged down, but the other signal remans relatively unchanged and just looses its sharp corner at the end of the waveform. The cursors can easily be moved to the start of the fault, the adjustment has both coarse and fine control to allow the cursor to be placed quickly and accurately. I can only have two vertical cursors on the screen at a time, as I am only interested join one point on this waveform, the second cursor is moved off screen, but the location label for this cursor and the differential calculation remain on the screen. It would be really nice if the cursor menu could include the selection of a single cursor and disable the other labels for this type of measurement.

 

So when faults are present it is easy to see both the traces, but when they aren't the traces overlap one another as seen below.

 

image

 

To show that both traces are present, the rotor reflectometer has a push button that will split the traces whilst pressed to give confidence that there are no faults. As you can see, this is easily captured by the oscilloscope. Using the Rigol, I would have had to use the reference waveform again and manually moved one of the traces using the vertical position function on the oscilloscope.

 

image

 

These traces are being captured due to the persistence function, that can be altered by tapping onto a blank part of the screen. The persistence mode always seems to default to 'auto' when the unit is switched on, so automatically captures the traces. You also have the option to turn it off if required.

 

 

{gallery} PScreen persistence functions

image

Screen persistence modes

image

Auto persistence options

image

Infinite persistence options

 

Using the auto function, the screen persistence does sometime produces noise, so does need to be turned off. If left in auto the traces come and go with the push button operation. If switched to 'infinite' mode the split trace is captured and remains there until the screen is cleared, or a new measurement started.

 

The alternative to using the screen persistence is to save the channel into one of the reference waveforms that can be displayed on the screen at the same time as the channel. The oscilloscope has the advantage of being able to carry out maths functions using the reference waveforms as well as the channels, so a subtracted waveform can then be displayed.

 

{gallery} Using the reference waveform

image

Reference waveform split from channel and maths channel added

image

Maths channel setting for waveform analysis

image

Analysis with inter-turn fault added to simulator

 

The reflectometer I have is an old style single channel unit, so this is the only way I have found to carry out an analysis on the scope. More modern reflectometers have a dual channel output and so the double transit trace can be split over two channels on the oscilloscope and then a live analysis carried out by subtracting channel 2 from channel 1. A new dual channel reflectometer costs close to this oscilloscope though, so I won't be seeing one of them for a while.

 

The following video shows the measurements above actual being carried out.

 

You don't have permission to edit metadata of this video.
Edit media
x
image
Upload Preview
image

 

Generator Test

 

The original plan was to take the rotor reflectometer and 3 Series MDO up to a generator and capture the tests on an actual generator rotor. Unfortunately, I am currently working on some switchgear issues at multiple sites, so this test has had to be delayed, probably until the new year. When I have had chance to complete this test, I will finalise this blog. Since I know that the scope will easily capture the traces from the tests already completed, this aspect was always really about how portable the unit was and how it would handle being away from the workbench.

 

Update 18th January 2020

 

This is the update from the actual test carried out on a generator rotor that is currently in preservation. As with all these assets that enter preservation, parts get borrowed for other installations and maintenance performed is minimal. As a consequence, a lot of the lights are out around the area, so the photography is not the best, the power sockets have also been removed, so I decided to test out the 3 Series MDO with a Power Cube to provide the 230V supply.

 

As the generator has quite a large bedplate, finding an area to setup the instrumentation is very easy.

 

image

 

With the rotor reflectometer and oscilloscope powered up, the power drawn from the supply was 83W as seen below. The cube has a 300W capacity, so it will quite happily run this test setup for some time.

 

image

 

The initial signal on the oscilloscope contained a high level of noise and was virtually unreadable. I assume that the noise is coming from the Power Cube and filtering through on the common earth from the cube to the oscilloscope, as the supply is effectively floating and is not connected to a solid earth. The instructions for the Power Cube, do contain a warning that noise may be present on televisions and media players.

 

image

 

I struggled to get the oscilloscope to trigger in a stable manner, even with average sampling turned on. I managed to capture a reading for the double transit trace, but could never get a reading for the single transit trace, due to the noise on the signal. I did manage to carry out the maths analysis with the reference signal and the 23 us transit time is similar to previous tests carried out.

 

image

 

image

 

The single transit signal above, is unreadable and the noise on the signal prevent me from taking any accurate measurements.

 

For comparison, I carried out the same test with the Riglo 1052E supplied from the Power Cube and it to showed a high level of noise on the signal.

 

{gallery} RSO Test with Rigol DS1052E

image

RSO Signal capture on Rigol DS1052E

image

32W Power Drawn by Rigol and Reflectometer

image

Screenshot of RSO Signal on Rigol

image

Low Pass Filter applied to RSO Signal by Rigol maths function

The one difference between the two oscilloscopes is that the Rigol does have a Low Pass filter function built into it, that I could use to get a more stable signal, than on the 3 Series MDO. However, once this is activated, a lot of functionality is then lost within the scope as the filter will be turned off as soon as any adjustments are made.

 

You can also see a huge difference in the quality of the screenshots between the 3 Series MDO and the Rigol, that will make a huge improvement to report writing in the future.

 

Whilst the test setup has worked, I need to do some more investigation and work to see if I can overcome the noise generated by the Power Cube. I did try the same test with the rotor reflectometer and its test delay away from site and a similar amount of noise was picked up. For some reason, when capturing the screenshots during this test, the 3 Series MDO screen flashed white and I lost a lot of detail in the screenshot as seen below.

 

image

 

I then altered the test setup, so that the oscilloscope was powered from the mains and the reflectometer from the Power Cube and could easily capture the RSO traces, so the noise only appears to be an issue when I power the oscilloscope from the Power Cube.

 

image

 

Conclusions

 

I think that the 3 Series has worked reasonably well in this application. Calibration checks went according to plan and the majority of the results were comparable to the usual method of using the 53220A53220A counter. Only being able to have four measurements across the two channels was a bit of a hinderance, it would be nice to be able to have a few more, especially if less than four channels are being utilised.

 

Capturing the traces of a test for a report were extremely easy, the size of the screen makes it much easier to see faults in comparison to the old Rigol oscilloscope. Measurements are easy to add and the coarse / fine control of the cursors makes them easy to position. For this particular test, it would be nice to have the option of just one cursor for fault marking and have the calculation and cursor B information boxes disappear from the screen if the B cursor is moved out of view, or have the option to turn it off.

 

Update 18th January 2020

 

The tests out on site were a little awkward, but this appears to be because I powered the oscilloscope from a Power Cube and not a direct mains supply. I will need to do some more work on this to see if I can find a resolution as the Power Cube is actually quite useful for power test apparatus. However the noise picked up occurred on any oscilloscope I used and not just the 3 Series MDO, so it is a global issue and not specific to the Tektronix oscilloscope.

  • Sign in to reply

Top Comments

  • three-phase
    three-phase over 5 years ago +3
    Found out what I had done to turn the screenshots into a white background. There is an 'ink saver' option on the save screen that I must have caught with one of my clumsy fingers and enabled. When ticked…
  • genebren
    genebren over 6 years ago +2
    Great update in your series on the Tektronix 3 series MDO. Interesting subject, reflectometry. Back in the 70's I did a lot of work on antenna systems, and the reflectometer was a super quick and easy…
  • DAB
    DAB over 5 years ago +2
    Very good update. Well explained. DAB
  • three-phase
    three-phase over 5 years ago

    Found out what I had done to turn the screenshots into a white background.

     

    There is an 'ink saver' option on the save screen that I must have caught with one of my clumsy fingers and enabled. When ticked, the screenshot has a white background.

     

    image

     

    When unticked, the black background is present again.

     

    image

    • Cancel
    • Vote Up +3 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • DAB
    DAB over 5 years ago

    Very good update.

     

    Well explained.

     

    DAB

    • Cancel
    • Vote Up +2 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • three-phase
    three-phase over 5 years ago

    I have added an update today to this blog to include the tests carried out on site on a generator rotor in preservation, and the 3 Series MDO with the rotor reflectometer.

     

    Kind regards

    • Cancel
    • Vote Up +1 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • three-phase
    three-phase over 6 years ago in reply to genebren

    Many thanks Gene. Yes reflectometry is very interesting, especially if you like your maths. This is a pretty basic use of the phenomenon and the test gear is very bespoke and only really useful for this particular test.

     

    Our main use for it would be for locating underground cable faults, but unlike your tests, this is done on an isolated cables, so we only need to know how far to walk before we starting digging a hole.

     

    Kind regards

    • Cancel
    • Vote Up +1 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • genebren
    genebren over 6 years ago

    Great update in your series on the Tektronix 3 series MDO.  Interesting subject, reflectometry. Back in the 70's I did a lot of work on antenna systems, and the reflectometer was a super quick and easy way to troubleshoot.  We had several antenna array system, with up to 16 elements, along with 100's of RF switches and trying to find the faulty element/switch/cable/connector was relatively easy.  We would attach the reflectometer to each of the antenna leads in turn and look for unexpected impeadance changes (open/short/leakage/etc.).  We could also manually control the various RF switches and verify proper impedances. We would typically have multiple failures after an intense electrical storm, causing bad readings or outright failures.  We would then travel out to the faulty towers and work out the issues.

    • Cancel
    • Vote Up +2 Vote Down
    • Sign in to reply
    • More
    • Cancel
element14 Community

element14 is the first online community specifically for engineers. Connect with your peers and get expert answers to your questions.

  • Members
  • Learn
  • Technologies
  • Challenges & Projects
  • Products
  • Store
  • About Us
  • Feedback & Support
  • FAQs
  • Terms of Use
  • Privacy Policy
  • Legal and Copyright Notices
  • Sitemap
  • Cookies

An Avnet Company © 2025 Premier Farnell Limited. All Rights Reserved.

Premier Farnell Ltd, registered in England and Wales (no 00876412), registered office: Farnell House, Forge Lane, Leeds LS12 2NE.

ICP 备案号 10220084.

Follow element14

  • X
  • Facebook
  • linkedin
  • YouTube