Tektronix TBS1102C Digital Oscilloscope - Review

View table of contents ...  

RoadTest: Tektronix TBS1102C Digital Oscilloscope

Author: geralds

Creation date:

Evaluation Type: Test Equipment

Did you receive all parts the manufacturer stated would be included in the package?: True

What other parts do you consider comparable to this product?: Oscilloscopes at entry class

What were the biggest problems encountered?: USB connection. Hardware also software. As well a problem I encountered was, if the USB connection was loosing the firmware freeze and the TBS1102C was rebooting. After booting, once or twice, the scope the firmware seemed to freeze and showed wrong signals. After set to the default setup by pushing the button "default setup" the firmware worked normally.

Detailed Review:

Greetings Tektronix, greetings Randall !

 

Thank you very much for letting me take part in this roadtest for testing the TBS1102C.

 

First impression:

The delivery was problem-free, all parts that are described in the delivery note were available.

After switching on the TBS1102C, booting took about 30 seconds until the start screen is displayed in a stable manner.

The last device and measurement settings are activated when the device is switched on again.

For better understanding, I switched first from the English to the German  language then I set the date and time, then I checked the firmware version.

Then I set the basic settings by pressing the button "Default Setup".

 

Starting the TBS1102C, initialization within 30 seconds during setup, warming up in the first minute.

 

1st measurement with the TBS1102C

image

TBS1102C_ starting the1st measurement (the probes were adjust before)

All of them worked well.

--

Application:

For the application, I used the oscilloscope's test signal and two development boards for simple tests.

imageimageimageimage

On the Microchip Curiosity-HPC I used the PWM generator and on the click board I simply used the 12MHz oscillator.

With the UNO and the Honeywell board I used the humidity sensor with it's I2C port.

This was enough for showing the applications with the TBS1102.

 

First thinking about:

What does the TBS1102C provide? The attached datasheet as well the manual are telling it us.

What is inside the TBS- scopes? The document 3GW_29364_0_HR.pdf shows some mainboards of such o-scopes.

 

What are the differences between some scope at this class?

Tektronix has made some comparisons.

 

imageimage

imageimage

imageimage

imageimage

imageimage

---

Everything first - please read the manuals for details as I will show everything in a global description.

Just like I'm going to tell what I found out about the features and what was missing; so that I don't do a doctorate dissertation.

 

Adjust the probes:

Since each device has its own impedance, nominal 50 Ohm, the probes must be adjusted on the built-in trimmers before use.

The TPP0100 probe reaches up to 100MHz and the BNC connector has a C-trimmer for setting the signal edges of a square wave signal.

As a comparison, I still used my old Hameg O-Scope and used its probe HZ51.

This probe, it reaches up to 150MHz, has two setting options.  The first is at the measuring tip, there is set for the LF edge.

The second is at the BNC connector, there is set for the HF edge (the corners of the square-wave signals will fine-tuned; if wrong adjusting also the amplitude will be manipulated).

More details can be found in the manual and in the help menu.

  TPP0100 - HZ51 Hameg HM1503 used for compensating the impedances of the probes and used for comparing the zoom functions.imageimageimage

How to compensating the impedance of the probes.

 

Compensation the probe properties to the scope is a standard exercise and always necessary. Finished.

The point I see here is that the test signal terminals are open and could be shorted.

- We know that current/voltage sources should always be designed as a socket.

Here with the TBS1102C, in contrast to the other devices, also from Tektronix devices.

As can be seen in comparison to the Hameg HM1503, these sockets are used, and the advantage is

that the shield line terminates directly at the probe tip and not with a loop wire, which affects the impedance in detail.

The second thought is, as can be seen in the comparison between the probes TPP0100 and HZ51:

The HZ51 has two setting options so that the square edges can be adjusted more precisely,

which is important for the HF setting (as can be seen when switching the test signal from 1kHz to 1MHz).

This is also a point I would like for the TBS1102C - a switchable square wave test signal.

 

Well, the compensation option for the TPP is in the BNC connector, which makes it easier -

"simply because the cable on the test object is an" extension "of the signal wire".

However, the probe cable always has a different impedance than the test object and it must therefore be adapted.

So I see the advantage of two to three setting options on the probe cable, which is also done with the corresponding

Tektronix HF probes in order to really reduce the repercussions of the supply cable, regardless of whether the measurement is low or high frequency.

So I would like to see at least two setting options for a 100MHz cable.

---

/Argument/ "it is just an EDU scope" --- NO, this argument does not count. "cheap, cheaper, cheapest" /or vice versa / - this is not a good argument.

 

Note:

I always think positively - promoting great products.

The TBS1102C is a good oscilloscope, but with limitations that I miss a lot.

I am a fan of Tektronix! The first time I came into contact with Tektronix equipment was in 1978.

---

What does the TBS1102C provide:

Key features include:

 

■ Bandwidth variants 50 MHz, 70 MHz, 100 MHz and 200 MHz

/ The bandwidths of the TBS 1000C series cannot be purchased as an option. Each device is manufactured with the defined bandwidth, which is then defined in its device name.

So, e.g. a TBS1102C, 100MHz with 2 measuring inputs, or TBS1202C, 200MHZ with 2 measuring inputs, are separate devices and must be purchased as such./

■ 2 channel models

■ Large 7 inch WVGA wide-screen colour display

■ Sample Rates of 1 GS/s on all Channels

■ Up to 20K points record length on every channel

■ Up to 5,000 waveforms/second capture rate

■ Supports 32 automated measurements

■ Edge, Runt and Pulse Width triggers

■ FFT analysis for waveform spectrum analysis

■ USB 2.0 Host ports for quick and easy storage of screen images, instrument settings, and waveforms to USB flash drives; installing firmware updates; and loading waveforms and settings from saved files

■ USB 2.0 Device port for direct PC control of the oscilloscope using TekVISA connectivity, and other remote connectivity tools that support USBTMC

■ Scope Intro provides a built-in overview of oscilloscope concepts and an introduction to the controls and features

■ Help Everywhere displays graphics and short text descriptions when you access the menus for most oscilloscope settings

■ Courseware function provides on-oscilloscope teaching instruction, with hundreds of courses available on the Tektronix Education Web page and the ability to easily create courses specific to your education needs

---

 

So let's go with some applications. I stepped thru the user manual.

I do not describe the buttons as well the front end, this will be described well in the manual step by step.

I just will show the most used measuring applications, also the "special features" of the TBS1102C.

 

HELP - The TBS1102C offers a excellent help under the HELP menu.

Under the help categories you can find:

- Course

- Scope intro

- Help Everywhere

 

You can find a very detailed description about the scope in / oscilloscope basics /, in / TBS1000C overview/, and in / TekSmartLab /.

In the course menu you will find fine tutorials, described in detail, with interactive instructions on how to prepare, set up and carry out measurements step by step.

imageimage

imageimage

imageimage

 

What I found out by walking thru:

Some of the description texts are printed double, in English as well in German (my preferred language, in other languages may be also double texts)

image

Walking thru the HELP menu:

 

Walking thru the UTILITY menu:

 

The menus working well.

----

APPLICATIONS:

With the Microchip Curiosity-HPC as well with the Arduino UNO combined with the Honeywell sensor shield I started the measurement series.

I show the individual measurements with corresponding videos and describe in text what I have found. Unfortunately my voice doesn't fit right now, sorry, so these videos are silent.

 

As shown above, after switching on and booting, the TBS1102C is ready to start within 30 seconds, in its basic position. But, the scope needs warming up about 30 minutes.

After a weekly or depending on the series of measurements, the probes should be checked and readjusted to compensation regularly.

You should also push the "DefaultSetup" button regularly to bring the device to its basic position. You can let the scope do the set measurement automatically with the "Autoset" button.

Autoset is a fast way to acquire and display a waveform.

Autoset automatically sets the trigger type to edge, sets the threshold level to 50% of the signal level, and analyses the input signal and adjusts the oscilloscope acquisition,

horizontal, and vertical settings to display five to six waveform cycles. Autoset is ideal for quickly displaying a repeating signal waveform.

If the measurement results are unsatisfactory, the measurements can be reworked using e.g. Trigger, Horizontal-Aquire, Vertical-channels, or with the onboard tools "Measurements".

The tools can be found under the menu "MEASURE" - simply push the button and it appears that menu for selecting the measurements.

 

The vertical channels can be labelled with pre-set labels, or by typing a free label text.

// It works fine. //

What I found out:

The multipurpose knob at scrolling thru the list as well scrolling thru the function list works very good.

But in that case after pushing this button for selecting the item, this knob runs very fast toward and a bit too "slippery" so that a pushing can occur at the next item.

In this case I would suggest making a mechanically gently stronger multi-purpose knob. Very light, just so that the scroll wheel cannot turn any further.

 

Walking thru "LABEL":

 

Measure:

 

---

Trigger:

 

Trigger Ch1 or Ch2, if the frequencies are close to each other or further away.

 

Cursor:

 

 

Automatic measurements:

When selecting the 32 automatic measurements, a total of 6 memory locations are available for the two vertical input channels.

So you are already limited if you have to make several measurements for both channels at the same time. Careful division of the measurement series is therefore important.

Now there are videos in which the automatic measurements are also switched on for the series of measurements.

The selection and setting of the applications worked without any problems; except for the multifunction button mentioned above, which was quite easy, rather too easy to turn.

 

I suggest that each vertical channel should have 6 memory locations.

So, I would suggest a firmware update.

In addition - the cursors should also be coupled with the measurement signals. With certain measurements this only works with the time / frequency cursors.

(e.g. FFT) So, it should be possible to switch all cursors to the signals.

 

Rising - edge:

 

Falling - edge:

 

ZOOM:

The zoom function is a very practical function, in this way you can see details more closely.

Compared to the analogue scope Hameg HM1503, two time bases are used here.

This enables a simultaneous display of the waveform on the screen, in normal view and zoomed view, as well as the view of the waveform position.

With the TBS1102C, zoom is simply activated with the "Zoom" button, the signal is then shown magnified by 10 times and the corresponding menu items are activated for this.

This function works well. In addition, two videos on how to set this.

 

  

 

Automatic measurement period with cursor - use with a PWM signal:

 

XY - display mode:

The XY display mode plots the signal amplitude of channel 1 (X) against the amplitude of channel 2 (Y), where channel 1 is the horizontal axis and channel 2 is the vertical axis.

XY display mode is useful for showing signal phase or frequency relationships between periodic signals. The resulting XY plots are called Lissajou patterns.

 

XY - 2 phase sine signal  - on the clock generator board measure the 12MHz oscillator signal

 

XY - sine moves to the ratio of the pulse width

 

FFT - tuneable sawtooth waveform

 

The Math function:

 

Application with the Arduino UNO and the Honeywell sensor shield with the humidity sensor

For the next measurements I used the serial interface with the I2C protocol.

The period, burst and RUNT trigger have been examined for these measurements.

 

I also used the Tektronix software OpenChoiceDesktop Appliction (TDSPCS1), Kickstart and TekScope.

 

imageimage

 

TBS1102C_I2C_Burst

 

TBS1102C_I2C_Periode-Burst:

 

I2C - RUNT trigger:

 

 

Software:

 

Tektronix offers fine, powerful analysis software for the TBS1102C.

These can be downloaded from this website:

https://de.tek.com/search?keywords=TBS1102C&type[0]=type:software

 

 

Tek - OpenChoiceDesktop Appliction (TDSPCS1):

 

Device selector image

 

"RUNT trigger" / image storingimage

 

"I2C protocol" / dataset storing image

 

This software work so far well. But it needs a lot of PC CPU time resources, so about over 20%. Hm... it is a fine image of the front-end, real view of the buttons, may be....

 

Kickstart:

 

image

 

image

This software is very easy to use. It work very well. You need a license (sorry, it's a bit expensive) to work with it.

 

TekScope:

TekScope is by far the most modern and flexible analysis software from Tektronix.

You can find more information via this link.

https://www.tek.com/software/tekscope-pc-analysis-software

 

However, before you can download this software you have to register.

After checking the user data (USA law, due to export regulations it will be checked accordingly, which takes about 2 working days) you can complete the registration and the download link will be activated.

After installing and downloading the appropriate license for this front end, you can use the functions for displaying the measured values saved by the measuring device.

You can still NOT! Establish a connection to the measuring device, you have to buy the appropriate licenses!

So - it is basically a picture book display program as long as no device license is installed!

 

For me that was a stressful situation for the time being, because normally I expected that at least one cable connection with the device would be visible.

 

So it really took two days to read everything on the website, to wait for the answer to my support request, and then to install the ordered license (you get a link to download the license file via email) on the PC.

Well... now I can test the trial version for 14 days.

Fortunately, I waited until the end of the last few days with testing via TekScope, otherwise no test series for this road test would have been possible. Uff...

 

So, for the serial protocol measure I need a one year subscription license (cost about 200 USD).

-> The TecScope PC Waveform Analysis Software contains ALL what you need for the supported equipment's. But you must pay for this every year in form of a license.

 

TekScope Base Package

TekScope PC Analysis Software

$0.00

 

free to use as image viewer

 

Multiscope analysis package

Multi-Scope Analysis

Remotely connect and acquire data directly from your oscilloscopes to view and analyze the data from all channels simultaneously on the same screen.

Analyze data from up to 32 channels and 4 scopes.

1yr $3,000

Perpetual $7,500

 

Jitter Measurements and Analysis Package

Jitter Measurements and Analysis

Advanced Jitter and Eye Analysis - Jitter, Eye, Amplitude, Time Measurements

1yr $1,200

Perpetual $2,900

 

Low Speed Protocol Package

Low Speed Protocol

I2C, I3C, SPI, RS-232, SPMI, I2S, LJ, RJ, TDM, CAN, CAN-FD, LIN, FlexRay, SENT, 100B-T1 Automotive Ethernet,

Mil-STD-1553, ARINC-429, Spacewire, USB 2.0, eUSB2, PSI5, SVID, 10 Base-T / 100 Base-TX Ethernet, MDIO, NRZ, 8b/10b, MIPI D-PHY, Manchester, SDLC, 1-Wire, MIPI C-PHY CSI/DSI

1yr $2,500

Perpetual $6,300

 

Power Integrity Analysis Package

Power Integrity Analysis

Digital Power Management and Analysis - Ripple Analysis, Transient Analysis, Power Sequence Analysis, Jitter Analysis

Power Management Serial Decode and Analysis (SPMI)

1yr $850

Perpetual $2,000

 

SpectrumView Analysis Package

SpectrumView Analysis

Any analog signal may be presented as a waveform, spectrum, or both.

Time domain waveforms and spectrum are synchronized.

Waveform and spectrum display may be independently adjusted.

Settings are made with spectrum analyzer controls such as Center Frequency, Span, and Resolution Bandwidth (RBW).

1yr $650

Perpetual $1,600

 

Power Electronics Package

Power Electronics Analysis

Advanced Power Analysis - Input Measurements, Amplitude Measurements, Timing Measurements, Switching Measurements, Output Measurements Magnetic Analysis - Inductance Measurements, Magnetic Property Plot, Magnetic Loss Measurements, Current vs. Integral of Voltage Plot

Inverter Motor Drive Analysis - Input Analysis, Ripple Analysis, Output Analysis

1yr $850

Perpetual $2,200

 

This license you need for the TBS1102C:

Remote Analysis for Bench Oscilloscopes Package

Remote Analysis for Bench Oscilloscopes

Remotely connect and acquire real-time data directly from your bench oscilloscope. Make analysis and decode all essential protocols - I2C, SPI, RS-232/422/485/UART, CAN, CAN-FD, LIN.

1yr $200

Perpetual $500

 

Greetings  Randall

At the beginning for this roadtest in Februar, I asked for this license.

Well paying every year makes the TBS1102C very expensive. I explained in summary.

If it possible, is there a way to get the perpetual license?

 

What have I encountered:

Please see the videos:

 

The application, measure the I2C protocol:

TekScope-TBS1102C_I2C waveform what I want to decodeimageimage

 

TekScope software hang again after waveform refresh - Problem USB connection TBS1102C and TekVISA

 

 

 

Because the USB connection between the PC and the TBS1102C no longer works properly, the TekScope can no longer be closed properly.

I had to stop the program using the task manager.

But the USB search continues. Tek-VISA also no longer displayed the connection. The TBS1102C suddenly restarted, so its firmware hung up.

 

 

After restart the TekScope and the TBS1102C //// (...waiting for the next crash?)

 

TekVISA - Problem checking the checkbox in the USB menu item.

image

 

 

Summary:

The automatic measurements I used with the TBS1102C worked fine.

The problems I found during the measurements are described above.

 

However, at times it happened that the "Background Brightness" utility in the utilities did not work correctly.

When scrolling with the multifunction knob to set the values, the counted value is sometimes not shown on the display. The value is already counted in the background, but it only becomes visible when the button is pressed.

If I get out of this menu item and then I go back to the brightness setting and set it again, then increment and decrement are actually displayed.

So it seems that some background programs do not work properly, so that they sometimes fail.

 

I had the biggest problems with the USB interface from the TBS1102C to the PC.

This actually led to the failure of the PC software and failure (hang up) of the oscilloscope firmware, so that the oscilloscope automatically rebooted.

 

The videos in connection with the TekScope software show these problems, total crashes.

I even restarted the PC to be on the safe side!

 

The TEK OCD software has its own problem:

At times it requires over 20% PC CPU resources on my PC. So, there seems to be some overhead in the background that is slowing the system down a lot.

The thick and colorful natural views of the software's front-end images probably consume a lot of software code.

 

TEK-Visa is probably a very outdated driver program because it keeps losing the USB port.

In addition, you cannot switch on the automatic USB search, you can just switch on the other interface searches.

 

USB:

If you disconnect the USB cable connection from the TBS1102C to the PC while it is active,

then the TekScope crash, the Tek-Visa looses the USB and the oscilloscope hang. The oscilloscope will then automatically reboot (watchdog?).

 

However, the TBS1102C is recommended, but with a few drawbacks.

The important thing for me here is:

It would really be an advantage if the protocols and decoders for the serial interface measurements are already built into the device.

It is clearly also a question of license costs.

But! Please compare the similar oscilloscopes as from Rigol and Keysight and others.

These already have this interface measurement built in as a standard and this devices cost much less than this TBS1102C.

Ok, some of them have not a spectrum analyser. But they have a touch screen as well as they have much analogue channels, more built in functions, then the TBS1000 series.

In the discussion on your website at TekScope, there is the question of whether the software license on the PC or in the device.

I would like to tell you right away: Install in the device!

Because what if the PC fails, the hard drive crashes and the connections are constantly lost, e.g. the USB (or LAN) here,

then this is extremely expensive and frustrating at the same time, because you have to constantly repair it.

Total cost can be "astronomic", all licenses are bounded to the host-ID!!! What if the harddisk crash? - All is done.

Because the software was crashing I had must restart my computer! This is a no-go for an application like TekScope.

The measurements require real-time measurements! - USB with the PC, connected to an external real-time equipment is by no means the solution.

 

I'm really a fan to the Tektronix as I mentioned above. This oscilloscope was my favorite what I worked with!

image

 

Maybe something was wrong with the installation of the TekScope but I think it wasn't, I checked it multiple times.

A hardware connection within the devices / software must be absolutely stable (yes, with the exception of the exceptions) because we believe this at this point and use it as a reference and error isolation.

Seen in this way, I can recommend the TBS1102C very well as a stand-alone device, if - if the "Serial interface protocol analysis" option is built into the device.

Then a source of error due to a relatively unstable USB (PC tasks, threads, are not realtime, ergo USB is not realtime) connection is eliminated.

 

I will use the TBS1102C as well! It has some fine features.

Ah yes, an idea:

At front there is the USB port for saving the datasets. - Idea: you can for example use this USB port as an input for an logical analyser, or not? Why not? It would be fine, if some extensions  are connected on the TBS.

The Function button on the TBS can be used for this advanced function, or not? - This key has only one task at the moment, to play coarse / fine for the setting wheel. If you give it more tasks, this button will definitely be happy.

 

Thank you very much, that a was selected to be a part of this roadtest!

It was a honour for me.

Best Regards

Gerald

---

Edit the Review, April 13.:

Greetings Tektronix, greetings Randall 

 

Now after I checked once more the USB also the software TekScope as well as the TekView it looks that the USB connection seems stabile.

One of the possible points. what is to do:

Close all active programs and unplug all unnecessary USB devices from the PC during installing the TekView and the TekScope.

Yesterday I disconnected the Arduino UNO and the TBS1102C from the USB, then I restarted the computer. The keyboard and the mouse stayed connected.

After that I checked the TekView - and yes, then it was possible to check the USB checkbox in the menu item "Search Criteria" .

And then I reconnected the o-scope and the Arduino UNO to the PC and well, all the time until now the USB stayed stabile; no more was breaking the connection.image

TekScope - aquire was succesful

image

 

But one point:

Every time the TekScope overwrites the settings in the TBS1102C with values that are stored in the TekScope by default but are not compatible with the TBS1000 series.

For example that where values for a GHz oscilloscope; giga samples or so by default. Also the probes by default are 1GHz probes.

- or the horizontal time: the TBS needs 1, 2, 4, 10; and so on... the TekScope does switch to other values per default.

 

So my suggestion is that beginners in particular should help:

When I choose a device like this TBS1000 series, it would be nice if I could already choose the device from a list of supported devices by default. The parameters for this device would then apply and match.

It is difficult for beginners to program from the beginning after installing the software.

If you already have such great software, it certainly isn't a big deal to include such a list of devices.

Querying the data via the USB connection is not easy, however, since this does not take place in real time, since the data packet keeps failing and a complete data packet was not always transmitted. Sometimes worked probably sometimes did not.

 

Best Regards

Gerald

---

Anonymous
  • Hi shabaz,

     

    The OpenChoice Desktop is free downloadable an it will be linked at the Tek Website where that instrument will be offered.

    This is not an analysis software just picture viewer.

     

    The TekScope software you can only download after registration at this web link, what I mentioned in the review:

    https://www.element14.com/community/external-link.jspa?url=https%3A%2F%2Fwww.tek.com%2Fsoftware%2Ftekscope-pc-analysis-s…

     

    The software itself is free to use without any license just as a  waveform viewer for reading stored images. But!..... -> see below..

    But it have all including, what you need to analyse with that supported instruments.

    image

    So you need a license for connecting to any of the supported instruments, you need still to buy it.

    And this licenses are very expensive, as I mentioned above in the review. "grrrr...."image

     

    The TBS1102C cost about 1100 Euro (Farnell), the necessary license /essential serial protocol analysis/ cost for 1yr. 200USD, permanent 500USD.

    Comparing to this Rigol, cost about 1200Euro (options included as permanent license in the instrument, bundle offer for a small time) :

    https://www.element14.com/community/external-link.jspa?url=https%3A%2F%2Fwww.amazon.de%2FRigol-Digital-Oszilloskop-Multi…

    It has 4 channels, touch screen, 16 ch. logic analyzer, FFT, automated measurements, free PC software, and much more...

     

    In my opinion, Tektronix should create software that is distinct and separate from the instruments, and it must contain everything that instrument needs.

    - A TBS software, a what ever software ....

    However, Tektronix built a "juggernaut" of software (over 928MB download) "All in One" that supports much larger instruments, the number of which are more than just the family of the TBS 1000 and TBS2000 series.

     

    But: image .. now after I read a bit deeper - License: TekScope is free, well, but that license expires after 1 year, then you will need to renew it.

    Oh, dear...

     

    Best Regards

    Gerald

    ---

  • Hi Gerald,

     

    I'd only used the OpenChoice Desktop, but I thought I'd try the TekScope software (even though it's unlikely my 'scope will support it : ( however I got nowhere with it.

    Maybe it's the wrong program, I could only see Download option for "TekScope Anywhere".

    I only see Help, About and Install Options, yet the help refers to clicking on a File menu which doesn't exist, and selecting Open.

    I guess I need a paid license to go further : (

    image

  • Thank you very much Gerald, you are giving me a lot of clues so that I can make a less impulsive decision.

    Now I use an inexpensive 8 channel logic analyzer capable of analyzing signals up to 6Mhz and an ADALM 1000. They are not enough to satisfy my curiosity and spy on some protocols.

    At a professional level I am dedicated to developing software for card payment solutions and I continually need to spy and debug serial communications, USB, bluetooth, NFC, RFID, Mifare, ISO / IEC 7816, etc. There I use commercial solutions that allow me to spy and debug communications. I always wonder if I couldn't go a little further and understand what goes on in the physical layer.

    But the real reason is that I am approaching retirement age and in the future I would like to pursue the passion that I had as a child, electronics. I am not clear if to create things or out of curiosity to understand how they work. I began my university studies in robotics and automatic control electronics, but since at that time there were hardly any programmers and they paid very well, I left the university to go to work and did not graduate, although I did complete the six-year degree and all the courses.

    Over the weekend I will read your review again to see if I find in myself what I want image

  • Hi !

     

    Many Thanks.

    To your question: "whether I regret buying the O-scope" ..

    Hm .... "Hobby" is an extremely broad area.

    Unfortunately, I can't say anything about it because I don't know your professional experience or that in the hobby area.

    There are also wishes, so how far do you want to go, where should the limits be.

    In the TBS1000 series, the maximum level is 2 analog inputs with up to 200MHz. - TBS1202C.

    The functions are the same in this series.

    There are no built-in options, except for the spectrum analyzer, which is rarely used and is more likely to be used in the audio sector or for analyzing distortions (up to half the oscilloscope input frequency).

     

    Example because I read your link to Arduino:

    If you want to program on the Arduino, then you will certainly have to do with various serial interfaces.

    The TBS can represent two serial signals with 2 inputs, but some require 3 or more signal inputs.

    These protocols can usually be added to oscilloscopes as a purchase option, which can cost a lot more than the TBS basic device. Unfortunately, and that is what I have touched on in detail above, you can only analyze these logs with the TekScope plus software, which you can purchase with a license ($ 200 annually, $ 500 permanent).

     

    Well, you looked at my road test, the USB interface is a problem. The logs are displayed, but the querying of the data is not real-time, the Scope is actually stopped when querying and the data is transferred to the PC and then displayed after a long time.

     

    So it is better if these protocol measurements are built into the device, because then there is no line in between that can provoke disturbances in the measurement data.

    But these devices already cost a lot! - But the competition from Tektronix now offers inexpensive devices, where more options are already built in as standard.

    For example from Rigol, or from Keysight.

    Those were just the essential interfaces. Then it goes on ... Where do you want to continue with your hobby?

    Where are the next borders?

     

    Unfortunately, Tektronix is slowly coming down to the "normal citizen", they are more at home in the electronics elite.

    In contrast, Keysight has already learned a lot and is already working quite diligently in the lower small to medium-sized company segment.

    Rigol is a real Chinese and was unpopular for a very long time because the devices weren't the most accurate for a long time.

    But now Rigol is very accurate and affordable with many features at the same price as the TBS1000.

    Yes, $ 1000 - Rigol already gives you a 4-channel 100MHz with the serial interface protocols as standard and touchscreen.

    The automatic measurements are similar to those from Tektronix.

    So, half to a quarter of the price, or twice or more functions for the same price as the desired device (TBS ...).

    The Tektronix TBS2000 series cost about double, so they are a bit expensive.

     

    Please look and compare, then choose for yourself the most reasonable / not the best, but good enough for your "whole" life.

    My tip for a better hobby: Rigol

    https://www.rigol.eu/products/oscillosopes/mso5000.html

    -> look at Amazon,,,,

    https://www.amazon.de/Rigol-Digital-Oszilloskop-Multi-Touch-Bildschirm/dp/B08TR5P5R9/ref=sr_1_11?dchild=1&hvadid=8088285…

     

    //// Sorry to Tektronix, I'm a fan of them.

    But I really want to speak my opinion honestly, never lie anywhere or say discriminate against anyone.

    I also want the best of the best and have to save wherever possible. And I tell what I made wrong if anywhere was wrong.

     

    Best regards

    Gerald

    ---

  • Hi Thanks for you review.

    I have received an award for a project valued at $ 1,000 and I have the opportunity to acquire this same oscilloscope model.

    I am a person who is mainly dedicated to software and I want to learn electronics little by little as a hobby.

    Will I regret it if I purchase it?

  • Hi Shabaz,

    Many Thanks!

    I keep looking at your posts in Element14, because I can really learn a lot from you.

     

    Yes, that's right, a lot has happened in the FFT area.

    Lately I've also been studying demos and manuals from competitor manufacturers.

    What is happening now on the market for devices for the "common man" is really fantastic, but difficult at the same time.

    You can find a lot comparable, because all manufacturers want to satisfy their customers. But on the customer side, it's even more difficult to separate the good from the normal.

    Unfortunately, there is already a trend towards "cheap", cheaper "... but this also means that quality is losing out.

    When I remember, the tenor was always: "What do you need that for?"

    "Only this or that is enough anyway .."

    That is now satisfied and therefore you can no longer really get to the top products, because these stand out from the elite, the global players and eat everything smaller.

     

    This Tektronix software TekScope is such an example of "you can do anything, but if you pay heavily".

    With this huge software, I would have built a list in which you choose your suitable device from a device list, and then the rest of the software should be locked away after selection. -> One device, -> one license, done.

    But for this selected device, all device functions should then be free and these so-called "options" should simply be available as a standard in the device and openly accessible.

     

    Here will shown the differences of licenses between the PC an the measurement device / Item: PC vs. Oscilloscope Licenses. 

    https://www.tek.com/software/tekscope-pc-analysis-software

     

    This TekScope can do nothing! Not really. This is my opinion.

    It is a picture book display software. It looks beautiful, and then?

    Measurements are made with the measuring device, but analysis should also be done with the device, not with such software.

    Real-time measurements are very delicate. How long has it been possible to transmit a dynamic process in real time via a USB interface over several meters of cable? Does not work at all!

     

    I have real problems with the acquisition and readout of the triggered HEX data that run via the target interface I2C.

    Every second the Arduino UNO asks the sensor for 5 data packets, 1 address and 4 data. These data of this humidity sensor are constantly changing, so the oscilloscope cannot freeze them.

    And then I randomly ask for this data with the push on a button. The connection is already hanging because something has changed between two query events. Therefore, even on the PC, which is several meters away, it is not possible to continuously query via USB. It just has to crash.

    Then the PC itself has to do a lot of tasks before that fetched data will be printed on screen.

    That is the problem what I see here.

    Of course, I may be wrong, because I am not Tektronix, nor did I develop this software. Tektronix sure knows why they do this.

     

    My mind, this software can only used for remote control and checking from time to time once a day or so, looking for if all is right. Or if you have to document something.

     

    Yes, even Thomas Edison was clever! To build a lightbulb with a filament that will burn out after 10,000 hours.

    It's similar today. The "filament" license burns out after a year.

     

    In the next few days I will no longer be able to play with the software because the trial license is expiring.

    But I don't want to spend $ 200 a year on a pocket radio at the moment.

    Then so I would say, for someone who has not yet: I prefer to save on a device that already has these functions built in and doesn't cost anything extra.

     

    Best Regards

    Gerald

    ---

  • Hi Michael,

    Thank you for your observation. Yes, I absolutely agree with you.

     

    Yesterday I found a possible "bug" - problematic point - (is always forgotten).

    During the installation you have to disconnect the unnecessary USB devices from the PC.

    Even then, before the installation, you should clean up the PC as far as possible, e.g. delete the temporary stuck files, and so on.

    Perhaps best to do after restarting the PC, because then the programs are more or less in their basic position.

    Not every PC is used in a variety of ways. PCs with many applications, for example development environments, that access the same resources would be more problematic.

    In addition, the PC should meet the basic requirements.

    Well, then of course a super company would like to design and sell its products as fantastic as possible. But they cannot consider ALL scenarios that exist on planet earth.

    Well, that's how it works or it doesn't work ... 50/50 chance ...

     

    Best Regards

    Gerald

    ---

  • ---

    Edit the Review, April 13.:

    Greetings Tektronix, greetings Randall

     

    Now after I checked once more the USB also the software TekScope as well as the TekView it looks that the USB connection seems stabile.

    One of the possible points. what is to do:

    Close all active programs and unplug all unnecessary USB devices from the PC during installing the TekView and the TekScope.

    Yesterday I disconnected the Arduino UNO and the TBS1102C from the USB, then I restarted the computer. The keyboard and the mouse stayed connected.

    After that I checked the TekView - and yes, then it was possible to check the USB checkbox in the menu item "Search Criteria" .

    And then I reconnected the o-scope and the Arduino UNO to the PC and well, all the time until now the USB stayed stabile; no more was breaking the connection.image

    TekScope - aquire was succesful

    image

     

    But one point:

    Every time the TekScope overwrites the settings in the TBS1102C with values that are stored in the TekScope by default but are not compatible with the TBS1000 series.

    For example that where values for a GHz oscilloscope; giga samples or so by default. Also the probes by default are 1GHz probes.

    - or the horizontal time: the TBS needs 1, 2, 4, 10; and so on... the TekScope does switch to other values per default.

     

    So my suggestion is that beginners in particular should help:

    When I choose a device like this TBS1000 series, it would be nice if I could already choose the device from a list of supported devices by default. The parameters for this device would then apply and match.

    It is difficult for beginners to program from the beginning after installing the software.

    If you already have such great software, it certainly isn't a big deal to include such a list of devices.

    Querying the data via the USB connection is not easy, however, since this does not take place in real time, since the data packet keeps failing and a complete data packet was not always transmitted. Sometimes worked probably sometimes did not.

     

    With this settings works mostly (not but all acquire):

    TBS1102C TekScope_ I2C - horizontal

     

    image

    TBS1102C TekScope_ I2C - horizontal

    image

     

    TBS1102C TekScope_ I2C - Port-Channels

    image

     

    If I find more then I'll telling here.

     

    Best Regards

    Gerald

    ---

  • Nice road test report.

     

    DAB

  • Hi Gerald,

     

    It was interesting to see the FFT video, looks like this area has improved compared to older 'scopes. I think the OpenChoice Desktop software could be improved in some areas too, but it works reliably and does it's main job well I guess!

    Great review!

  • Comprehensive review Gerald - thanks.

     

    Your work on this highlights the need to look carefully at the software support and licensing on modern scopes and the need to include this in evaluating the total value for money and lifetime cost proposition.

     

    MK

  • Hi,

    Yes, this feature is nice.

    I don't know that there is in all models implemented. I think this features are in EDU models of such oscope.

    In the bigger one oscopes I think there are other, more informative help files.

    Please look for at the Tektronix website what is included what not.

    What I know, Tektronix make all the best for students and service engineers and they publish one of most detailed manuals what I've ever read.

    In the service manuals what I have, are extremely clear descriptions in detail.

  • The instruction/tutorial page within scope is really nice feature! Do thery come in all models of Tek scopes?