Test & Tools
Test & Tools
Blog DC Voltage Measurement Accuracy: AD584KH voltage reference
  • Date Created: 20 Jun 2023 8:47 PM Date Created
  • Views 5224 views
  • Likes 14 likes
  • Comments 37 comments
  • dmm6500
  • reference
  • keithley
  • AD584
20 Jun 2023
Follow up on  DC Voltage Measurement Accuracy: Measuring DC voltage with high accuracy on a DMM6500 .
I bought a little voltage reference based on Analog Device's AD584 .
The reference is the AD584KH. The kit supports the 4 programmable reference values. It comes with a label showing measurements on a HP 3458A
image source: analog devices datasheet
Calibration values (25° / 49% humidity):
  • 2.49974
  • 5.00203
  • 7.50113
  • 10.00277
Calibrated by http://www.haoqielec.com/wap/en/aboutus.html
Checking Keithley DMM6500
I let the reference and DMM warm up for an hour. Room temperature was 26°. Humidity unknown
I used Keithley's TSP script "Measuring DC voltage with high accuracy". I used that in the previous post too.
The Test Script Builder log gives a higher resolution than the 6500 display.
I let the reference settle for 5 minutes on each value before running the script and getting the measurement
CAL (HP 3458A DMM6500 round
2.49974 2.4997806153 2.49978
5.00203 5.0021082424 5.00211
7.50113 7.5012037198 7.50120
10.00277 10.002870735 10.00287
Thoughts and remarks? Post them in the comments
    shabaz over 3 years ago in reply to ggabe +2
    Hopefully these photos can be zoomed up without the content system compression causing issues: The marking is A1 HB , or possibly A1sHB , and it looks like it is Si2301 (P-ch MOSFET) . I have some…
  • ggabe
    ggabe over 3 years ago in reply to Jan Cumps +1
    Could it be a ground current? What if you battery power and also connect the USB ground?
  • scottiebabe
    scottiebabe over 3 years ago in reply to Jan Cumps
    When I saw 31 digits I thought you just mashed the keypad, but no it was output from a calculation. Does MS calc use quad floats?
    Jan Cumps over 3 years ago in reply to scottiebabe
    (back to reality) The effective number of bits in this measurement, analysed by Renesas software, was 16 bits. (in line with the decimal devaluation of this discussion, they report 16,6014260934955 bits Slight smile). The datasheet reports the possible resolution of 24-bit delta-sigma A/D converter as: "Up to 23-bit effective resolution"
    calculator is in scientific mode, who am I to argue Yum
  • scottiebabe
    2^(-24) = 60 e-9, that's a lot of extra digits from oversampling and averaging...
    Hardware is limited by the software: