Test & Tools
Test & Tools
Blog DC Voltage Measurement Accuracy: AD584KH voltage reference
  • Date Created: 20 Jun 2023 8:47 PM Date Created
  • Views 5231 views
  • Likes 14 likes
  • Comments 37 comments
  • dmm6500
  • reference
  • keithley
  • AD584
20 Jun 2023
Follow up on  DC Voltage Measurement Accuracy: Measuring DC voltage with high accuracy on a DMM6500 .
I bought a little voltage reference based on Analog Device's AD584 .
The reference is the AD584KH. The kit supports the 4 programmable reference values. It comes with a label showing measurements on a HP 3458A
image source: analog devices datasheet
Calibration values (25° / 49% humidity):
  • 2.49974
  • 5.00203
  • 7.50113
  • 10.00277
Calibrated by http://www.haoqielec.com/wap/en/aboutus.html
Checking Keithley DMM6500
I let the reference and DMM warm up for an hour. Room temperature was 26°. Humidity unknown
I used Keithley's TSP script "Measuring DC voltage with high accuracy". I used that in the previous post too.
The Test Script Builder log gives a higher resolution than the 6500 display.
I let the reference settle for 5 minutes on each value before running the script and getting the measurement
CAL (HP 3458A DMM6500 round
2.49974 2.4997806153 2.49978
5.00203 5.0021082424 5.00211
7.50113 7.5012037198 7.50120
10.00277 10.002870735 10.00287
Thoughts and remarks? Post them in the comments
    shabaz over 3 years ago in reply to ggabe +2
    Hopefully these photos can be zoomed up without the content system compression causing issues: The marking is A1 HB , or possibly A1sHB , and it looks like it is Si2301 (P-ch MOSFET) . I have some…
  • ggabe
    ggabe over 3 years ago in reply to Jan Cumps +1
    Could it be a ground current? What if you battery power and also connect the USB ground?
    DAB over 3 years ago
    Nice post Jan.
    I would say your accuracy is pretty good unless you are in a laboratory setting that requires more precision.
    Jan Cumps over 3 years ago
    A sweet little Brymen BM235
    2.499
    4.999
    7.50
    10.00
    One day later: The same measurements, again with meter and reference warmed up. It's 1° warmer than yesterday.
    2.4997878455
    5.0021127197
    7.5012086365
    10.002881856
    Jan Cumps over 3 years ago in reply to Jan Cumps
    measurements with 121GW:
    CAL (HP 3458A)        121GW
    2.49974                      2.5002
    5.00203                      5.0031
    7.50113                      7.499 (range switches here)
    10.00277                    10.000
    hehee