The Labfacility L300 Pt100 temperature alarm / on-off controller can be used in conjunction with a PC to provide accurate monitoring and alarm or on-off control of up to 8- zones simultaneously. It can also be used as a stand-alone instrument without the need for a PC. The PC software supplied with the instrument allows control, configuration, measurement, logging, charting, alarm & relay configuration and calibration functions via a PC. The in-built self-calibration facility of Pt100 ranges is simple and uses plug-in precision resistors.
L300PT
Farnell order number: 2770798
L300PT - Controller, Temperature Alarm, On - Off, PT100, 8 Zones, USB InterfaceL300PT - Controller, Temperature Alarm, On - Off, PT100, 8 Zones, USB Interface
General DescriptionBuilt-in display for selected channel or all channels auto-scrolling. PC software included for remote control and measure, logging, configuration and calibration. 8 x configurable change-over relays 10A/250V. Simple operation. On-board memory 512 sets of readings
Input / RangesPt100 to IEC751, 3 wire -200°C to 850°C
Linearisation Conformity±0.05°C
User InterfaceFront panel keys for selecting channel number for display or auto-scan selection; front panel keys for relay configuration and alarm parameters. 8 x LED indicators for relay actuation
Self CalibrationUser facility incorporated. The instrument auto-calibrates on every A/D cycle
Power Supply6Vdc (5.5-9.0V) via universal mains adaptor (supplied) 120-250V 50/60Hz
AccessoriesSupplied as standard on CD-ROM Remote control & measure: - Log readings to file Download to PC Logging, charting, alarm configuration and calibration
Accuracy
better than ±0.05°C ±0.1% of range
Applications
Food preparation - Storage facilities - Technical educational establishments - Environmental - R&D - Heating & ventilation setup - Building & energy management - Instrumentation laboratories -
Experimentation - Refrigeration/freezer plant monitoring - Museums and Galleries
  • Low cost – high performance
  • USB PC interface
  • 8 Pt100 (3 wire) inputs
  • Built-in display for selected channel or all channels auto-scrolling
  • Resolution 0.1°C on display, 0.01°C in software
  • Self-calibration feature
  • Select °C / °F
  • PC software included for remote control and measure, logging, configuration and calibrationL300PT Back View
  • 8 x configurable change-over relays 10A/250V
  • Simple operation
  • CE marked
  • RoHS compliant
Specifications
Specification at an ambient temperature of 20°C Measurement
Input / Ranges
Pt100 to IEC751,           3 wire      -200°C to 850°C Note: all inputs are non-isolated and sensors must be of insulated construction.
Accuracy
Linearisation                     ±0.05°C
Pt100                               range better than ±0.05°C ±0.1% of range
Zero drift                          ±0.01 % of span per °C
Span drift                         ±0.01 % of span per °C
Display                             LCD, backlight
Display resolution            Pt100 range 0.01°C
Indication                         Channel No., measured temperature
                                        (°C or °F)
Sensor open circuit          Upscale indication
detection & indication
Ambient operating           0 to 50°C
Temperature
Alarm/Control
Alarm modes                High / Low / Band/ control
Relay contacts              x3 normally open */common/ normally closed * The contact position when the relay is de-energised Rated 1 0A/250V, resistive load. Relays ( 1to 8) can be assigned to any input and polarity (normal or Inverse) selected
User interface               Front panel keys for selecting channel number for display or auto-scan selection; front panel keys for relay configuration and alarm parameters. 8 x LED indicators for relay actuation.
Storage temperature      -20°C to 70°C
Display                          LCD with backlight
Input Terminations         8 x Pt100, terminal blocks
Relay Terminations       4 x 6-way connector
PC Interface                 USB
Power supply                6Vdc (5.5-9.0V) via universal mains adaptor (supplied) 1 20-250V 50/60Hz
Logging interval           5 seconds to 1 hour
PC software                Supplied as standard on CD-ROM Remote control & measure: - Log readings to file - Download to PC - Logging, charting, alarm configuration and calibration
Standard accessories The L300 is supplied with a power supply adaptor, USB lead, PC software, and instruction manual (on CD).
imageL300 Datasheet.pdf