Hi,
Does anyone know what input voltage TENMA 72-1051072-10510 can handle Cannot seem to find that info on farnell or on google
Hi,
Does anyone know what input voltage TENMA 72-1051072-10510 can handle Cannot seem to find that info on farnell or on google
Are you looking for the maximum voltage you can measure? Or the maximum voltage the oscilloscope can tolerate before there is damage? For the maximum measurable voltage, looking at the max volts/div is the most appropriate method.
For maximum ratings of the scope input and probe, both are available in the manual/datasheet.
Max Scope Input: PDF page 51. It can tolerate up to 400 Volts-peak.
Max Probe Input: PDF page 61, shows that in the x1 setting max input voltage is 150 V and in the x10 setting it is 300 V.
Are you looking for the maximum voltage you can measure? Or the maximum voltage the oscilloscope can tolerate before there is damage? For the maximum measurable voltage, looking at the max volts/div is the most appropriate method.
For maximum ratings of the scope input and probe, both are available in the manual/datasheet.
Max Scope Input: PDF page 51. It can tolerate up to 400 Volts-peak.
Max Probe Input: PDF page 61, shows that in the x1 setting max input voltage is 150 V and in the x10 setting it is 300 V.