Dear friends
Few months ago I decide to built a multi channel pressure box, so I bought 16 Honeywell ASDX differential analog pressure sensor (Model: ASDX RRX005NDAA5). the datasheet of this type of sensor attached.
Base on datasheet, this sensor has 3 active pin, pin No.1 is for voltage supply, Pin No. 2 is Analog output and Pin No. 3 is GND.
I design a PCB board to mount all 16 sensors on a single board. I use a 5Vdc power supply device for PCB board, now when I TURN ON the device and there is no applied pressure on sensors (just ambient pressure), the output voltages are almost 2.47 - 2.49 Volt (I checked Pin 1 voltage for each sensors and all of them is same 5Vdc). the problems are here:
Problems:
1. I expect when there is no applied pressure, the output voltage be 2.5V but as i mentioned above it is a little less, what is the reason?
2. when I applied an specific pressure (for example 300Pa) most sensors shows different voltage in their output and so it is impossible to calculate the Pressures. the datasheet said i can calculate pressure by below Eq.
3. by the changing of the pressure, change in output voltage of sensors are not equal. for example if I increase the Pressure 500Pa, the voltage change for one sensor is almost 1.5V but the voltage change for another sensor is 0.7V!!
I will be so appreciate if help me