element14 Community
element14 Community
    Register Log In
  • Site
  • Search
  • Log In Register
  • Community Hub
    Community Hub
    • What's New on element14
    • Feedback and Support
    • Benefits of Membership
    • Personal Blogs
    • Members Area
    • Achievement Levels
  • Learn
    Learn
    • Ask an Expert
    • eBooks
    • element14 presents
    • Learning Center
    • Tech Spotlight
    • STEM Academy
    • Webinars, Training and Events
    • Learning Groups
  • Technologies
    Technologies
    • 3D Printing
    • FPGA
    • Industrial Automation
    • Internet of Things
    • Power & Energy
    • Sensors
    • Technology Groups
  • Challenges & Projects
    Challenges & Projects
    • Design Challenges
    • element14 presents Projects
    • Project14
    • Arduino Projects
    • Raspberry Pi Projects
    • Project Groups
  • Products & Partners
    Products & Partners
    • Arduino
    • Avnet & Tria Boards Community
    • Dev Tools
    • Manufacturers
    • Multicomp Pro
    • Product Groups
    • Raspberry Pi
    • RoadTests & Reviews
  • About Us
    About the element14 Community
  • Store
    Store
    • Visit Your Store
    • Choose another store...
      • Europe
      •  Austria (German)
      •  Belgium (Dutch, French)
      •  Bulgaria (Bulgarian)
      •  Czech Republic (Czech)
      •  Denmark (Danish)
      •  Estonia (Estonian)
      •  Finland (Finnish)
      •  France (French)
      •  Germany (German)
      •  Hungary (Hungarian)
      •  Ireland
      •  Israel
      •  Italy (Italian)
      •  Latvia (Latvian)
      •  
      •  Lithuania (Lithuanian)
      •  Netherlands (Dutch)
      •  Norway (Norwegian)
      •  Poland (Polish)
      •  Portugal (Portuguese)
      •  Romania (Romanian)
      •  Russia (Russian)
      •  Slovakia (Slovak)
      •  Slovenia (Slovenian)
      •  Spain (Spanish)
      •  Sweden (Swedish)
      •  Switzerland(German, French)
      •  Turkey (Turkish)
      •  United Kingdom
      • Asia Pacific
      •  Australia
      •  China
      •  Hong Kong
      •  India
      •  Japan
      •  Korea (Korean)
      •  Malaysia
      •  New Zealand
      •  Philippines
      •  Singapore
      •  Taiwan
      •  Thailand (Thai)
      •  Vietnam
      • Americas
      •  Brazil (Portuguese)
      •  Canada
      •  Mexico (Spanish)
      •  United States
      Can't find the country/region you're looking for? Visit our export site or find a local distributor.
  • Translate
  • Profile
  • Settings
Test & Tools
  • Technologies
  • More
Test & Tools
Blog Practical use of the Agilent 34461A
  • Blog
  • Forum
  • Documents
  • Files
  • Members
  • Mentions
  • Sub-Groups
  • Tags
  • More
  • Cancel
  • New
Join Test & Tools to participate - click to join for free!
  • Share
  • More
  • Cancel
Group Actions
  • Group RSS
  • More
  • Cancel
Engagement
  • Author Author: 6thimage
  • Date Created: 29 Nov 2013 10:40 PM Date Created
  • Views 3471 views
  • Likes 0 likes
  • Comments 8 comments
  • RoadTest
  • 34461a
  • agilent
Related
Recommended

Practical use of the Agilent 34461A

6thimage
6thimage
29 Nov 2013

I was originally planning to show the use of the multimeter in a tutorial style, however, the multimeter is very easy to use and I feel that a tutorial approach is likely to seem, and be, patronising. So this blog post is just going to focus on practical measurements that can be done with the 34461A.

Microcontroller current consumption

The power consumption of modern microcontrollers can vary quite a lot, with many of them stating that they have very low power modes. I thought that it would be quite interesting to measure the power of an ARM Cortex-M4 microcontroller, as I have quite a few of the same development board - the STM32F4 discovery board.

The STM32F4 has four main power modes - run, sleep, standby and stop - with the discovery board having a specific jumper designed to measure the microcontroller’s power consumption.

The first mode I tested was sleep. The microcontroller was programmed, then held in reset for the first 30 seconds of the recorded data. It was then released from reset, where the programmed software would immediately place it into sleep. After another 30 seconds, the microcontroller was woken up, by pressing a button, with the microcontroller illuminating four LEDs to indicate that it is awake.

The image below shows the trend chart on the multimeter.


imageimage

The image above is a plot of the data, with annotations showing the different modes and the average current draw for each mode (with averaged 1-year errors).

The same process was done for the standby mode.

imageimage

The eagled eyed readers will spot that both the sleep and standby modes were measured with the range set to 100 mA, not to auto. Even though the reset would be best measured in the 10 mA range and the standby mode in the 1 mA range, it was observed that the microcontroller experienced brownouts when using autoranging. This was particularly noticeable with the standby mode, where the current draw crosses two of the multimeter’s ranges very quickly. Despite this, the 6.5 digit resolution of the multimeter still gives an excellent level of precision.

When the STM32F4 is placed into its stop mode, it has to be woken by its RTC. To save time, I decided that the device shouldn’t wake up for this test. So the images below show the STM32F4 being held in reset for 30 seconds and then 30 seconds of the microcontroller after it has placed itself into its stop mode.

imageimage

I was surprised to see that this had a higher current draw than the standby mode, so I think that I might not have disabled everything that is necessary to reduce the power consumption further. However, as we are concerned with the multimeter and not the microcontroller, we can see that the 34461A is very good at measuring the current consumption of a reasonably high-end microcontroller.

Resistor characterisation

I was originally planning to characterise a pack of resistors using the probe hold mode, however after finding out that the probe hold mode doesn’t support statistics I decided to do it by manual triggering. For this I used the 2 wire resistance mode, with 100 PLC and the SMT grabbers to hold onto the ends of each of the 96 resistors (which were still on their bandolier). I would have liked to have used the 4 wire resistance mode, however it was not practically possible as it introduced too much noise (as I was holding the other two probes). As the resistors I was measuring were 5.1 kOhm, the nulling of the leads was more than adequate for accurate readings.

The graph below shows two histograms, one generated by the multimeter and one generated in the graphing software from the collected readings.

image

Unfortunately, I managed to discover a bug in the multimeter firmware, where the histogram on the multimeter’s display does not update when triggering has been stopped. So I was not able to capture the screen with the correct histogram.

The table below shows the measured resistances with their 1 year accuracy.

Minimum

5073.58317 Ohm +/- 507.458 mOhm

Mean

5084.78991 Ohm +/- 52.1167 mOhm

Maximum

5105.37517 Ohm +/- 510.638 mOhm

With the following table showing how far these values are away from 5.1 kOhm.

Minimum

0.517977 % +/- 0.00995016 %

Mean

0.298237 % +/- 0.00102190 %

Maximum

0.105395 % +/- 0.0100125 %

As you can see from the table, all of the resistors fall well within the 1% specification.

Characterisation of voltage references

There are many applications where accurate and stable voltage references are required, for example in ADC circuits. So I was interested to see how well you can use the 34461A to characterise different voltage references.

MAX6126-30

The first reference I characterised was one of the highest accuracy references I could get my hands on - Maxim Integrated’s MAX6126. This is a +/- 0.02% reference with a temperature coefficient of less than 3 ppm/ deg C, so should ideally be quite stable. The 34461A was set to auto-trigger and auto-range, with 10 PLC, and was left for around an hour and a half to warm up. Just for the sake of intrigue, I left the multimeter recording for this warm-up time, with the following images being the results.


imageimageimage

Whilst practically not being very useful, it is quite interesting to see, from the trend graphs, the warm-up characteristics of the system - whether this is the multimeter or the voltage reference, is something that I have not looked into, as both had power applied to them at the same time.

The multimeter was then left for 9 hours measuring the voltage reference at the default rate (once every 200 ms). With the following images being taken from the multimeter.

imageimageimage

One of the things that I particularly like about the multimeter is its programmability, which means that the user can directly access the histogram data from a computer. Whilst I will cover this in a later blog post, I thought I should mention that the histograms I obtained from the multimeter was done over the LAN interface using a Python script on my computer.

The first graph shows the exact same data as what is in the previous image, just with a slightly higher amount of detail.

image

The following image shows the same data again (on the same x-axis) but with a log frequency axis, to show the infrequent readings with a greater detail. Additionally this graph has three vertical lines, which, from left to right, are the 1st quartile, the mean and the the 3rd quartile. This essentially shows that half of the data is contained within the two bins enclosed by the outermost red lines.

image

As the measurements are temperature sensitive, it also makes sense to show a graph of the temperature for the same 9 hour period. For almost a year now, I have had a raspberry pi collecting temperature measurements on a per minute basis. This uses an Analog Devices' ADT7410, which has an accuracy of +/- 0.5 deg C. The following graph shows a histogram of the temperature over the 9 hours, with the red lines indicating the same as before.


image

As the 34461A records the minimum, maximum and mean, we do not have to infer this from the histograms, with the values for 80,156 measurements being,

Minimum

2.99911586 V +/- 154.969 uV

Mean

2.99949370 V +/- 547.414 nV

Maximum

2.99951127 V +/- 154.983 uV

with the errors being calculated from the 1 year accuracies in the datasheet. From this, the range of the measurements is 395.41 uV +/- 219.169 uV.

As this is the 3 V MAX6126, we can calculate how far off the measurements are from the ideal reference value.

Minimum

0.0294713 % +/- 0.0051656 %

Mean

0.0168767 % +/- 0.0000182471 %

Maximum

0.0162910 % +/- 0.0051661 %

With the voltage reference having an accuracy specification of 0.02%, we can see that the minimum value measured is worse than what it should be, however this does not mean that the reference is operating outside its specification. On average the reference is within its 0.02% specification with a value of approximately 0.0169%, this can be trusted to a high degree due to the number of readings that have been averaged. Although this value will have some level of computational error in it, as the 34461A uses floating point maths for calculating statistics.

The first thing to consider is whether the accuracy is reduced due to the temperature of the reference. The temperature, measured by the ADT7410, had a minimum of 19.007812 deg C +/- 0.5 deg C and a maximum of 20.101562 deg C +/- 0.5 deg C. As the MAX6126 is specified for 25 deg C, the maximum temperature difference is almost 6 degrees. This affects the accuracy of the reference by 0.00179766 % +/- 0.000150000 %. With all the errors adding together to reduce the figure, the lowest the minimum value could possibly be is 0.0223580 %, which is still outside the specification.

It is also possible that the accuracy has been diminished by another factor. From the 34461A's tend graph, there are spurious negative spikes that were quite unexpected. To get the voltage reference as physically close to the temperature sensor as possible, the reference and multimeter were placed in close proximity to a wifi router, which may have added additional noise to the measurement. Additionally, as the voltage reference was hand soldered, it is possible that the accuracy of the reference might have been decreased due to thermal shock (i.e. too high of a temperature), however, I would have expected this to consistently decrease the accuracy of the sensor, rather than produce the observed spurious spikes.

MAX6126-21

The same was done with the 2.048 V version of the voltage reference.

imageimage

The histogram, as shown on the multimeter, has a much sharper peak than that of the 3 V version. And the trend plot, yet again, shows the spurious negative spikes, which suggests that it might be a noise issue.

The graphs below show the histogram data, with the second graph having a logarithmic frequency scale.

imageimage

As with the 3 V reference, the temperature was recorded and is displayed in the histogram below.

image

The table below show the data gained from the multimeter’s 80,156 readings, with the errors being calculated from the 1 year accuracies.

Minimum

2.04669303 V +/- 121.634 uV

Mean

2.04827951 V +/- 429.824 nV

Maximum

2.04831946 V +/- 121.691 uV

Much like the 3 V version, the voltage reference is within its specifications for the mean and maximum,

Minimum

0.0638169 % +/- 0.00593917 %

Mean

0.0136479 % +/- 0.0000209875 %

Maximum

0.0155986 % +/- 0.00594195 %

but not for the minimum reading. Once again, the error caused by the temperature is small, at 0.00179063 % +/- 0.000150000 %, and does not explain the minimum reading being, at best, 0.0559371 %, when its specification is 0.02 %. This minimum was taken from one of the seemingly random negative spikes that can be seen in the trend chart, and yet again, we can only speculate on its origin. However, it is possible to say that, on average, the voltage reference is within its specification.

ADR5040

The Analog Devices’ ADR5040, unlike the MAX6126, is a shunt voltage reference. The version I used came in a SOT-23 package and has an accuracy of +/- 0.2 %, with a temperature coefficient of 100 ppm/ deg C. Like for the MAX6126, the 34461A was set to auto-trigger and auto-range, with 10 PLC. The multimeter was left for longer to warm up (over 3 hours), however, this should make no difference to the measurements.

The first image shows the histogram on the multimeter,

image

you can immediately see that the peak is not as sharp. With the multimeter helpfully saying that the peak represents 29.9 % of the data, opposed to the 32.4 % and 60.7 % of the MAX6126. Whilst it is tempting to comment on the accuracies of the references based on this peak, it would be meaningless to do so as different bin widths were used for the different references.

image

The trend chart is fairly flat, but once again shows the negative spikes, which suggests even more that they are due to interference.

As was done before, the graphs below show the histogram plotted on a computer, with both non-logarithmic and logarithmic frequency.

imageimage

The temperature was recorded over the same period and is displayed in the histogram below.

image

Over the 8 hours, 71,241 readings were recorded. The minimum, maximum and mean of these readings is shown in the table below, again with the calculated 1 year errors.

Minimum

2.04377 V +/- 121.532 uV

Mean

2.04801 V +/- 455.897 nV

Maximum

2.04810 V +/- 121.684 uV

The percentage accuracy for these values is shown in the table below.

Minimum

0.206543 % +/- 0.00593418 %

Mean

0.000488281 % +/- 0.0000222606 %

Maximum

0.00488281 % +/- 0.00594158 %

The mean and maximum values are well within the specification of 0.2 %, surprisingly more so than for the more precise MAX6126. The minimum value is also within the specification when the temperature coefficient is considered, as at 100 ppm this equates to 0.0691406 % +/- 0.00500000 %. If we ignore the negative spikes, as they are likely to be due to interference, and take a low value from the histogram of 2.04785 V, we get an accuracy of approximately 0.00732 %. Overall, I think this is likely to be a very lucky case for this voltage reference and definitely not what can be expected in general.

Characterisation of current sink

I thought it would be quite interesting to measure and characterise a current sink, as a simple test of the current measuring capabilities against a reasonably known value. I decided to go with a fairly simple circuit,


image

using a voltage reference as one of the inputs to an op-amp. With this circuit, the current through the load is dependant on the 10kOhm resistor (R2), that is I_load = V_ref/R2. With the ADR5040 being 2.048 V, this makes the load current 204.8 uA.

Unfortunately, the resistor used for R2 was a standard 5% and so affected the load current. With the resistance being measured, in 4 wire mode, as 9.836040 kOhm +/- 1.08360 mOhm, the load current would ideally be 208.214 uA +/- 416.428 nA. This is, of course, treating the MOSFET as ideal and not considering temperature coefficients of the voltage reference or resistor. Additionally, this circuit was made on a breadboard and so this will also affect the accuracy of the current sink.

The current sink was measured for a little over 5 hours, using 10 PLC and autoranging. The following images show the multimeter’s histogram and trend plot.

imageimage

 

The trend plot here does not show the same negative spikes as was seen for the voltage references. This is likely due to this measurement being taken further away from the router, that was suspected of interfering with the readings.

The following graph is the histogram plotted in more detail.

image

And this is a histogram of the temperature.

image

Overall, the current sink worked reasonably well,

Minimum

208.104384 uA +/- 164.052 nA

Mean

208.126662 uA +/- 764.475 pA

Maximum

208.144134 uA +/- 164.072 nA

with a range of 39.750 nA, which is around +/- 0.01 % of the mean value.

Conclusion

I have found, from using the 34461A for these measurements, that the histogram feature is incredibly useful, largely because the bin data can be taken off of the multimeter for processing on a computer. The trend chart is quite useful, especially as it is not limited to the reading memory, as it allows you to roughly see what went on and diagnose potential problems (such as the interference with the voltage references). However, the histogram data purely on the multimeter is nice to look at but not particularly functional. This is because you cannot find the heights of the bins, unless you use one of the remote interfaces. To me, this is a pretty big oversight by Agilent and I would love to see them include a moveable cursor, much like the ones on digital oscilloscopes, so that you can use the data the multimeter has collected, on the multimeter.

  • Sign in to reply

Top Comments

  • robrenz
    robrenz over 12 years ago in reply to morgaine +1
    In the first part of my Keithley 2450 SMU http://www.element14.com/community/roadTestReviews/1603 review I made an attempt to address some of the negatives you correctly mention about videos. I put the…
  • morgaine
    morgaine over 12 years ago in reply to robrenz

    Robin, I'm very impressed that you took my little list of problems with video to heart and addressed some of them.  Many people just don't care about video effectiveness or educational merit.  You cared.  Thank you!

     

    I checked your inner video links, and your use of "?t=<time>" arguments is working fine with Youtube in Firefox.  It works despite the barriers I have put in place in the form of NoScript and FlashBlock add-ons for Firefox, which require me to provide an extra click before the video can start.  So, inner linking to video is a fully working mechanism, at least on Youtube.  Your yellow overlay section titles/headings also add metadata to the video itself, which is a very good idea.

     

    I've also watched the whole video as a continuous unit and it's very effective and well presented and produced --- well done.  I haven't yet examined the whole article though.  When I do, I'll add comments to your review.

     

    For now, full marks for a very good video presentation with due attention to making the video format a little more educationally effective than usual.  Well done!

     

    Morgaine.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • robrenz
    robrenz over 12 years ago in reply to morgaine

    In the first part of my Keithley 2450 SMU http://www.element14.com/community/roadTestReviews/1603 review I made an attempt to address some of the negatives you correctly mention about videos. I put the video on YT and then put links in the review to particular time indexed segments of the video that applied to that part of the review. I also included a index of the entire video so you could jump to any topic segment from the index.  I am by no means implying I did a good job but I wanted to make an attempt to address the negatives of videos.

     

    Robin

    • Cancel
    • Vote Up +1 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • morgaine
    morgaine over 12 years ago in reply to 6thimage

    Ian Griffiths wrote:

    Michael Kellett wrote:

     

    And congrats on a complete absense of videos !!

    I am not sure if that is sarcastic or not! Whilst doing this roadtest I have recorded a couple of things - like the boot time compared to another meter - but I don't know what I could have put in a video for this blog post.

     

    Although this is a rather unusual place to be discussing this topic, it is an interesting one.  I'll try to put a few objective bones on the subject, because just expressing an opinion about whether one likes or dislikes video personally isn't really very helpful.

     

    Video has these basic PRO/CON features, in no particular order:

     

    • CON: It's a monolithic lump.  Although there is timecode by which a particular section can be referenced, such references are not normally part of video URLs.  Also, the functionality that implements this may not be present at the client or server end, and so manual intervention is normally required to find an inner reference.  Unsurprisingly then, video today is generally considered a monolithic lump.
    • CON: It's not searchable.  Information is only useful if you can find it, and information contained within a video cannot generally (at the current state of technology) be located without a human physically watching the video.  Consequently, unless the video is accompanied by external descriptions or metadata, its information content is generally unknowable in advance of viewing.
    • CON: It's a time waster because there is no way to consume video other than in real time.  There is no means of obtaining a summary, of looking at the table of contents, nor skimming through it in any usable way.  The lack of inner references and structure contributes to the time wasting because you cannot easily limit what you end up viewing.
    • CON: The tools available for viewing video are generally poor for interactive use, typically providing only rudimentary media transport controls.  Pacing content to match rate of understanding is not automatic like it is with written material.  Browsers are especially poor video clients in this respect --- they're mostly designed to serve the needs of entertainment.
    • CON: Slowing down video for pacing purposes normally makes embedded audio unintelligible because phrase processing / beat slicing (as done with sampled music to preserve vocal content) is very rare in video.
    • CON: No cross-referencing.  Video content cannot easily reference earlier (or accompanying) material in the way that text typically cross-refers to tables and diagrams, and poor navigation controls exacerbate this problem.  As a result, the usual way for videos to reference earlier material is by referring to the memory of it.  This is very primitive.
    • CON: Video information reuse is almost impossible without medium conversion.  We have become accustomed to cut'n'paste as a means of rapidly selecting partial information, followed by simple reuse and rapid dissemination, but this is currently not possible with video for at least two different types of reason.  First of all, technology has not yet provided an easy cut'n'paste mechanism for selecting part of a video, and secondly, there are legal restrictions that strongly discourage video reuse through the instrument of copyright infringement, far more so than with textual material where quoting for fair use is normal and even expected.  As a result of these problems, the only simple and safe way to reuse the information contained in a video is to convert it into a different form --- for example, writing down some cited information (essentially making a new work).  This strongly hampers information reuse.
    • PRO: Watching video is a pleasant mind-numbing passive activity.  Almost everybody seems to enjoy vegetating in front of a screen some of the time --- even myself, which is why I removed TV from the house entirely.  It might provide useful relaxation, and probably activates the healthy biorhythms of sleep.  Reducing stress and tension is medically very helpful.
    • PRO: It requires no analytical skills to watch and enjoy.
    • CON: It requires good analytical skills to watch and understand video content.  It is one of the incorrect claims about video that the medium makes education, information and training easier to acquire.  This is mostly an illusion stemming from the pleasant nature of passive consumption.  Unless the mind is engaged very actively in the process, almost nothing is really learned (memory is not learning), and falsehoods are easily spread and acquired.  Learning through video is as difficult as through any other medium, and it is made harder when the tools support paced navigation only poorly.
    • PRO: From a lifetime of experience, we are good at watching people do things, and so we can learn quickly from observing solutions evolve dynamically on the screen.  When video material is of a type that can make use of this natural ability, it can be very effective.

     

    No doubt there's a lot more that could be said on the subject, but I'm conscious that this is a poor place for the discussion, so I'll end with just a quick opinion on how I think video fits in here.  The points enumerated above suggest to me that video is one of the poorest engineering tools we have available to us.  Although powerful in principle, our current state of technology does not deliver on the promise.

     

    Although video is the darling of social networking and a bedrock of the Internet's chattering classes, for engineering purposes I recommend sticking with less cool but more useful forms of presentation.

     

    Morgaine.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • 6thimage
    6thimage over 12 years ago in reply to michaelkellett

    I must have missed your comments - I know that the community has mixed views on videos, with a lot of people liking to make them and less people fully watching them. I think they can be useful when its done in the right way - I like watching Dave's EEVBlog on youtube, for example - but I have come across a few videos where they contain very little, both in content and knowledge, and it especially doesn't help when the quality of the video is low (i.e. camera bouncing everywhere, hard to hear audio, etc.). I think videos are quite difficult to do right, as you only have a few attempts at getting it right, whereas with text you can go over it several times in a week to get it right (each of these blog posts has taken me at least 2 weeks to write, and there is still more content that I want to add to them). I might try to find that video on the Wolfson board just to see how bad it is - I wonder if it is as bad as some of the youtube videos.

     

    I'm glad that you liked it - I was a little concerned that it might contain too many images that might not make sense to everyone. And also that people might not understand why the 1st and 3rd quartiles are important and why the error is drastically reduced for the mean values (or how to calculate the propagation of errors in general). I was hoping that if anyone didn't understand anything they would comment and say so, but I'm not sure if anyone would. But, as you say, if anyone wants to skip that, it is easier to do so with text.

     

    Thanks for your comments,

    Ian

     

    P.S. - I hope asking if your comment was sarcastic didn't offend you.

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • michaelkellett
    michaelkellett over 12 years ago in reply to 6thimage

    Absoulutely no sarcasm - if you had seen my comments about videos in (many) other posts you would know that I find them a waste of time. There is a prime example on E14 to do with the Wolfson Pi audio board - two blokes chatting for several minutes and conveying  a good deal less information in several minutes than you get across in a page of A4.

    I was really pleased that you had done diagrams and tables and words - that way I can study the bits that are interesting to me and skip the bits I've understood/ dont' care about etc.

    So thanks again, and remember, when it comes to videos, less is more and none is best !!

     

    MK

    • Cancel
    • Vote Up 0 Vote Down
    • Sign in to reply
    • More
    • Cancel
>
element14 Community

element14 is the first online community specifically for engineers. Connect with your peers and get expert answers to your questions.

  • Members
  • Learn
  • Technologies
  • Challenges & Projects
  • Products
  • Store
  • About Us
  • Feedback & Support
  • FAQs
  • Terms of Use
  • Privacy Policy
  • Legal and Copyright Notices
  • Sitemap
  • Cookies

An Avnet Company © 2026 Premier Farnell Limited. All Rights Reserved.

Premier Farnell Ltd, registered in England and Wales (no 00876412), registered office: Farnell House, Forge Lane, Leeds LS12 2NE.

Follow element14

  • X
  • Facebook
  • linkedin
  • YouTube