I purchased MAX6675. According to the datasheet, it said  "It accepts the input from a type-K thermocouple and converts the temperature to 12-bit data."
It is however, I purchase thermocoupler type-T instead of type-K, due to some mistake. Is type-T and type-K compatible to each other. From what I know, the temperature range, and sensitivity is different. It is confusing me that, MAX6675 is only for type-K thermocouple. Why is it?
    michaelkellett over 11 years ago
    The MAX6675 is full of cleverness to compensate for the scale factor and cold junction temperature of a K type thermocouple and output the digitized temperature in degrees. You could use a T type thermocouple and re-compensate external to the MAX6675 but it is a not a very good idea. You'll need a good understanding of thermocouples and the chip to do it. You would do much better to exchange the T type for a K type.
    dougw over 11 years ago
    Thermocouples have slightly non-linear response which may be compensated by chips like the MAX6675, but T and K have a pretty similar curve shape.
    K thermocouples change approximately 41 uV/degC while T thermocouples change about 43 uV/degC, so the error when using a T instead of a K is about 5%. As Michael points out - not great, but you may get some use out of it.