John Wiltrout's Blog Reverse Engineering integral DC Converters
  • Author Author: jw0752
  • Date Created: 6 Mar 2016 11:46 PM Date Created
  • Views 3061 views
  • Comments 24 comments
  • dc-dc converter
  • reverse_engineering
jw0752
jw0752
6 Mar 2016
The other day while trouble shooting a control panel from a Pelton Crane Delta Q Dental Sterilizer I noticed that two of the required voltages were not being supplied by the main power supply but were instead being synthesized by a small group of components on one end of the circuit board. The display on the sterilizer was totally dark and the problem was going to be in one of 2 boards. The ultimate decision we faced was which of the two boards costing $1000. and $600 respectively was bad. Unfortunately they would not be returnable if we guessed wrong and ordered the wrong board. Through a process of elimination I was able to determine that the problem was in some of the OEM components on the more expensive board. When I had gone as far as I could towards fixing the problem we broke down and ordered a new board.
My interest had however been captivated by the small group of components on one end of the board that was responsible for taking  +12 and  + 5 volt input supplies and outputting 24 volts and 40 volts to be used in the OLED display circuitry. Here is a picture of the board of interest and a closeup of the section responsible for the voltage conversion. At this time I had not yet realized that I was dealing with two distinct converters.
Over View of Display driver board
Closeup of 24V and 40V synthesis section
Back side of the board. This is a two layer board fortunately.
My first step to reverse engineering the circuit was to identify each of the components and to label a layout of the board. Fortunately I found that none of the components were OEM and that all were common and available in the shop parts supply. Where appropriate to understanding the circuit, Data Sheets were accessed. Here is the parts layout with labels.
The next step in the reverse engineering process is to create a schematic. This involves getting the component symbols on paper and determining their interconnections. For me this means a lot of time flipping the circuit board back and forth from front to back, visually tracing the the traces and vias, and using the ohm meter to verify connections. After a while, a rough draft of the schematic is produced by hand. A close inspection of the schematic and application of circuit logic often forces me back to the board to make sense of the circuit and to correct obvious gaps.
While it is possible to work off a crude schematic like this I always like to clean up the work and make a CAD schematic. Obviously, I got into a little trouble in the lower center of the hand produced schematic and we will clean up this section in the CAD production. Drawing freehand while reverse engineering a board usually leads to a messy first draft. Here is a CAD rendition.
With the schematic in hand it became clear that we actually had two converters on this board working off a single astable multivibrator. The 555 timer in this case is providing a 67 kHz 12 volt square wave. The section of the circuit that produces the 24 volt output uses this signal to drive a push pull set of NPN PNP general purpose transistors. The transistors in turn drive a voltage doubler with a 24 volt output. The 40 volt output is produce by driving an N Ch MOSFET with the 555 signal. The input voltage for this section is only 5 volts. When the MOSFET is in saturation energy is stored in the 100 uH inductor L1. When the MOSFET turns off this energy is released through diode D4 and stored in capacitor C4. Since the impedance is quite high the voltage is also quite high. Resistor R6 limits the current to Zener diode D5 which serves as a voltage regulator. Output in this case 40 volts is taken from the point in the circuit marked P1(7). C5 helps to decouple transients and noise from the output. C10 provides a reserve of current to L1 when the MOSFET turns on.
It is interesting to note that this design has been used by the sterilizer company for over a decade. When the board was first designed a 56 volt supply was needed and the requisite Zener diode was used to produce this voltage. As the technology of the OLEDs has improved the current required voltage became 35 volts and the only change to the circuit was the use of a 35 volt Zener. My rendition of the circuit creates 40 volts but this is only because my parts bins only had (2) 20 volt Zeners in this higher voltage range so I put them in series to approximate the actual circuit.
Now that we have the schematic the fun begins as we will bread board the circuit from the schematic to see if it will work as hoped. Here is a picture of the circuit bread boarded. Fortunately, except for the Zener, I had all the correct value components in the bins so I did not have to place an order.
For my initial test I have supplied +12 volts to the main rail and +5 volts to the MOSFET section of the circuit. I have an oscilloscope monitoring the output of the 555 astable multivibrator and I have DVMs on the 24 volt and 40 volt outputs. This is very unusual for me but it actually worked the first time. No Smoke, No Fire!  Here are the outputs as displayed by the test instruments.
Sorry for the poor image. I was too lazy to export the image and just took a picture of it.
This is my 40 volt output point. Since I am using 2 Zeners in series and probably do not have them in their optimal operation mA range I feel the 39.23 V is acceptable.
The regulation of this output is tied to the regulation of the 12 volt input of the circuit and I am very pleased with how close it came to 24 volts. The circuits that are using these voltage supplies are not precision and can tolerate a range of voltage.
The last thing that I did was to load the 40 volt circuit with a 10 K resistor and measured the voltage drop. The 10 K load produced a 2.39 Volt drop. Calculations indicated that the supply has an approximate 650 ohm ESR. At least 475 ohms of this ESR is the series current limiting resistor. I have left the test jig set up on the bench and will play with it for a couple days before tear down. I found this exercise helpful as I can see application of the concepts used in these 2 voltage converters will be useful in future designs.
    mcb1 over 10 years ago in reply to jc2048 +3
    I hope we're not encouraginging anyone to design outside the datasheet - a cardinal rule of professional design is that you always work within the parameters set by the manufacturer. Haha .... unless you…
    shabaz over 10 years ago +2
    Hi John, Excellent work as usual, and very relevant to using modern displays such as OLEDs. All the ones I've encountered need some particular high voltage, so it is good to see the various boost converter…
    jw0752 over 10 years ago in reply to Jan Cumps +2
    Hi Jan, I have run some experiments with "interesting" results. I modified the circuit so that it was more like your ideal by removing the 40 volt zener diode. The Vg = 5.1 volts, According to the Scope…
    Jan Cumps over 10 years ago
    Great great work, jw0752 !!
    The 40 volt part has a classic boost converter in it.
    I've redrawn your schema a little bit to make that more apparent.
    Here's the classic drawing (taken from Analysis of Four DC-DC Converters in Equilibrium )
    Jan Cumps over 10 years ago in reply to Jan Cumps
    In theory, the output voltage (before R6) should be:
    V = Vg / (1-D)
    Where V = the output voltage
    Vg the supply voltage (5V)
    D the duty cycle (you could check the duty cycle of the NE555 output with your Rigol's Measure functions, let's say it's 90% here)
    -> V =5V / (1 - 0.9) = 50V
    It would be cool if you could check that duty cycle going into Q9's gate, and the voltage before R6 to see if the equation holds.
    jw0752 over 10 years ago in reply to jc2048
    Hi Jon,
    Thank you for your insight about the continuous and discontinuous operation. I have recently read about this condition several times in different articles and this experiment is giving me a chance to see the effects. Also it was good to learn more about the body diode of the MOSFET. Tonight I plan to run an experiment where I will put a large capacitance only on the output of the converter. This causes the voltage to slowly climb towards the 129 volts. I want to observe at what point the current draw goes from 110 mA to the 250 mA. I also want to see if it is a relatively abrupt transition or if it is a gradual one. I was not aware that the body diode would avalanche like a Zener when too much reverse voltage is applied. I did not know where the extra power evidenced by the increase in mA was going but the body diode wasn't on my list of suspects.
    Thanks John
    Jan Cumps over 10 years ago in reply to jw0752
    John,
    a pure capacitive load is the same as no load in this application, because it's DC.  It will only play a role in the very first seconds while the converter is going into steady state (edit: that will not be reached because of no DC current path to stabilise on).
    I'd put a resistive load on there.
    You are of course correct but in this case the current that the converter can generate is small enough so with a suitably high capacitance it will be a minute or two before the voltage gets too high. This will give me time to determine the point at which the current goes from 110 mA to 250 mA and also whether the transition is abrupt or gradual. I want to reference the transition point back to the data sheet to see if there is a correspondence to the avalanche point of the internal body diode of the IRF511. Keep an eye on me though, I have been known to make my share of smoke and fire.
    mcb1 over 10 years ago in reply to jw0752
    Oven gloves as Personal Protection Equipment ... nice
  • jc2048
    jc2048 over 10 years ago in reply to jw0752
    I'll throw in a warning here that you are going to end up with a large capacitor charged to 130V with no real load to discharge it quickly afterwards. I know that you don't need the warning because you used to work with vacuum tubes, but others might not be so aware.
    An alternative to doing the experiment slowly, would be to do it quickly using your scope. If you put a low value resistor (less than an ohm - perhaps start with 0.1) between the source of the FET and ground, the voltage across the resistor would be proportional to the current (and one end is ground so it makes for an easy measurement with the scope). During the time that the FET was on you'd see the current ramp up as the coil established its magnetic field and poured energy into it. Normally, when the FET turned off the current would fall to zero because the coil's energy passes to the output capacitor through the diode, but if the FET diode breaks down you'd see the current ramp down again because the path would now be through the FET.
  • jc2048
    Hi Jon,
    Excellent warning about high voltage on capacitor to everyone including old guys like me. I also like your idea to monitor the current in a source resistor using the scope. I am getting more interesting tests out of this circuit than I originally thought I would.
  • jc2048
    jc2048 over 10 years ago in reply to jc2048
    I was curious about the FET body diode characteristic, so I thought I'd try it myself.
    Here's the circuit I ended up with after a bit of fiddling about. The first coil I tried was saturating [the ramp up of the current curved as it went up rather than being straight], but I found one that was ok up to the kind of currents you were trying. I didn't bother with the diode and the capacitor; I just establish the magnetic field in the coil and then discharge all of the energy through the FET. I used a IRF510, so the results will be a bit different to what you were seeing, but the principle is the same. The gate pulse comes from a function generator which lets me vary the current by changing the width of the pulse. [I was confident that this would work ok and not blow up the FET because the datasheet shows a similar circuit used for determining the energy the FET can cope with when switching an inductive load.]
    Here's the circuit on a piece of PCB material. For simple things like this, I prefer to "bird's nest" them rather than use prototyping blocks [I make a crude layout using a craft knife to cut V grooves in the copper]. It's a technique that can prototype dc-dc convertors quite well if you lay out the high current stuff as you would on a pcb and then bird's nest the sensitive low-current connections in the air above - not as good as a proper PCB, obviously, but far superior to stripboard or prototyping block [only works if the device has leads that you can see and solder to, though]. It's not obvious from the photo, but the current sense resistor is 5 x 1R resistors in parallel.
    The following trace shows the drain voltage [yellow trace] and the current [blue trace]. The 0.2ohm resistor in the source means that 20mV is equivalent to 100mA.
    The FET is on where the current ramps up to 200mA. When it turns off, the drain voltage leaps up to about 110V and the current declines as the coil's energy discharges through the FET. The discharge path must be through the FET because that's all that's left in my circuit and in any case we are measuring the current through the FET.
    This next trace shows the period when the FET diode breaks down in more detail. It's a pretty hard characteristic - it's not until right at the end, when the current is really low, that you see the curve of the "knee".
    This is the same, but with the current waveform turned off so you can see the voltage in better detail.
    Hi Jon,
    Great job. This is exactly what really helps. Thank you for taking the time to post your results for us to see. I won't thank you for going to all the work as I think you are like me and doing the work was a compulsion to try it for yourself. Sometimes I just can't wait to bread board a circuit or an Idea so that I can see for myself how the reality of the circuit will compare with the ideal of the circuit. It is really great to have you looking over my shoulder.
  • jc2048
    Thought it was about time I joined you all and did something, rather than just commenting.
    I've been thinking about the results and I think I was wrong to suggest that the curve was essentially the diode characteristic. Certainly the plateau up at 110V is determined by the avalanching of the diode in the MOSFET, but the curve back down is more to do with the parasitic capacitance (the capacitance from drain to source and from drain to gate within the MOSFET, plus the capacitance of the scope probe that's measuring the voltage, plus any other stray capacitance) discharging through the coil.
