element14 Community
element14 Community
    Register Log In
  • Site
  • Search
  • Log In Register
  • Community Hub
    Community Hub
    • What's New on element14
    • Feedback and Support
    • Benefits of Membership
    • Personal Blogs
    • Members Area
    • Achievement Levels
  • Learn
    Learn
    • Ask an Expert
    • eBooks
    • element14 presents
    • Learning Center
    • Tech Spotlight
    • STEM Academy
    • Webinars, Training and Events
    • Learning Groups
  • Technologies
    Technologies
    • 3D Printing
    • FPGA
    • Industrial Automation
    • Internet of Things
    • Power & Energy
    • Sensors
    • Technology Groups
  • Challenges & Projects
    Challenges & Projects
    • Design Challenges
    • element14 presents Projects
    • Project14
    • Arduino Projects
    • Raspberry Pi Projects
    • Project Groups
  • Products & Partners
    Products & Partners
    • Arduino
    • Avnet & Tria Boards Community
    • Dev Tools
    • Manufacturers
    • Multicomp Pro
    • Product Groups
    • Raspberry Pi
    • RoadTests & Reviews
  • About Us
    About the element14 Community
  • Store
    Store
    • Visit Your Store
    • Choose another store...
      • Europe
      •  Austria (German)
      •  Belgium (Dutch, French)
      •  Bulgaria (Bulgarian)
      •  Czech Republic (Czech)
      •  Denmark (Danish)
      •  Estonia (Estonian)
      •  Finland (Finnish)
      •  France (French)
      •  Germany (German)
      •  Hungary (Hungarian)
      •  Ireland
      •  Israel
      •  Italy (Italian)
      •  Latvia (Latvian)
      •  
      •  Lithuania (Lithuanian)
      •  Netherlands (Dutch)
      •  Norway (Norwegian)
      •  Poland (Polish)
      •  Portugal (Portuguese)
      •  Romania (Romanian)
      •  Russia (Russian)
      •  Slovakia (Slovak)
      •  Slovenia (Slovenian)
      •  Spain (Spanish)
      •  Sweden (Swedish)
      •  Switzerland(German, French)
      •  Turkey (Turkish)
      •  United Kingdom
      • Asia Pacific
      •  Australia
      •  China
      •  Hong Kong
      •  India
      •  Japan
      •  Korea (Korean)
      •  Malaysia
      •  New Zealand
      •  Philippines
      •  Singapore
      •  Taiwan
      •  Thailand (Thai)
      •  Vietnam
      • Americas
      •  Brazil (Portuguese)
      •  Canada
      •  Mexico (Spanish)
      •  United States
      Can't find the country/region you're looking for? Visit our export site or find a local distributor.
  • Translate
  • Profile
  • Settings
Test & Tools
  • Technologies
  • More
Test & Tools
Blog Tenma Model 72-2685 Digital Controlled DC Power Supply - First Impressions
  • Blog
  • Forum
  • Documents
  • Files
  • Members
  • Mentions
  • Sub-Groups
  • Tags
  • More
  • Cancel
  • New
Join Test & Tools to participate - click to join for free!
  • Share
  • More
  • Cancel
Group Actions
  • Group RSS
  • More
  • Cancel
Engagement
  • Author Author: fmilburn
  • Date Created: 26 Oct 2019 4:23 AM Date Created
  • Views 11330 views
  • Likes 13 likes
  • Comments 13 comments
  • product review
  • test equipment
  • tenma 72-2685
  • dc power supply
  • bench psu
  • psu
  • review
  • test hardware
Related
Recommended

Tenma Model 72-2685 Digital Controlled DC Power Supply - First Impressions

fmilburn
fmilburn
26 Oct 2019

Thanks to Project14 and the Electronic Toys contest I have a new bench power supply!  Thanks tariq.ahmad.  In this blog I will outline why I selected it and my first impressions.  I also do a quick out of the box check to make sure things seem to work as they should.

image

 

I have needed to upgrade my power supply for quite some time and have been on the lookout for something with the following specifications:

  • Reasonably priced
  • 0-30 V
  • 0-3 A
  • mA setting resolution and low mA load accuracy
  • Dual channel
  • USB / LAN interface

 

Dual channel would be nice as I occasionally need a second rail for either op amps or 3V3 and 5V but it was of lower priority.  In a similar manner I would like to be able to control the unit from a LAN but it is for the time lower priority also.  The mA setting resolution and accuracy are difficult to meet in a low cost unit and twiddling with the potentiometers on the one I have (fine and coarse) make it all but impossible.  Recently I noted that the Tenma 72-2685 Tenma 72-2685 was on sale for $68.02 at Newark (although it appears to be selling for $119.36 as I write this).  Here are the specifications - note the Current Load Regulation, Setup Resolution, and Setup Accuracy along with encoder control:

image

Unfortunately it is not dual channel and it can't be controlled by the computer but those are wants rather than needs.  So I snagged one :-).

 

Unboxing

 

It came nicely packaged and undamaged in a box which I shall keep in case I need to transport it.  The instructions are terse but sufficient.  My unit has a North American 110 V mains power cord (it is switchable to 220V).  It does not come with leads.

image

 

Quick Look Around

 

This is a basic power supply so our tour won't take too long.

image

Front:

  • The digit display is bright and easily read from all angles.
  • There are bright LEDs for Constant Voltage, Constant Current, Lock, and Over Current Protection
  • The encoder controls for current and voltage are great!  They allow precise setting and are reasonably quick to set once familiarity is gained
  • There is a power on / off button (blue)  which completely turns the unit on and off
  • There are ground, positive and negative terminals that accept banana jacks - there are no binding posts

Sides:

  • Ventilation
  • Four Phillips screws on each side to hold case together

Back:

  • Fan exhaust (runs only when needed)
  • Power socket and switch for 110 or 220 V
  • Fuse accessible from outside with a spare fuse
  • Approx. 1.1 meter power cord supplied

Top:

  • Handle

Bottom:

  • Four rubber feet

 

Overall it looks good and well built from the outside.

 

A Peek Inside

 

I am not qualified to assess a mains powered unit but popped the cover and had a quick look around.  I make no judgement other than the following:

image

  • Inside is clean, and things generally neat and tidy
  • Cable and wire is held securely and out of the way so as to not get pinched
  • Mains were grounded to the case all the way around as was ground terminal

image

  • The power MOSFETs are on a PCB that back to a large heat sink attached to the fan
  • Solder joints and connections look OK with no flux residue

image

  • Large Jicon capacitor has silicon to hold it in place, rated at 105 C and placed away from MOSFETs
  • Silicon used to hold some connectors and wire in place

image

  • The front panel PCB was clean and solder joints looked good.
  • What I assumed to be vias and cooling traces was interesting.
  • One largish capacitor on its side with no silicon to hold it down
  • The terminals did not look to be of high quality
  • No bodges or repairs seen anywhere

 

Trying it Out

 

I don't have my electronic load working yet and I don't plan to look at ripple or reaction to load changes so this is going to be quick and dirty.  I may come back for a deeper investigation in future but for now here is a quick look at the things that most interested me.

 

The transformer taps are at 7V, 14V, and 21V.  The specifications give the voltage range to be 0-30 V but it can be adjusted up to 31 V.  I hooked up a 10K resistor (measured to be 10.01 k) in series just to see how accurate the voltage setting and readout was at low currents.  The voltage was checked with a Tenma 72-102072-1020 bench digital multimeter (DMM1) and the current with an Aneng AN8008 (DMM2).  I have an assortment of inexpensive multimeter with the 72-102072-1020 being the best.  There are no calibration certificates but they are all in close agreement with better resolution than the Power Supply Unit (PSU).  Voltage was measured at the terminals so as to avoid influence of resistance in the wire leads. 

image

The results are in the following table.

 

 

PSU Setting (V)PSU Reading (V)DMM1 Reading (V)PSU Reading (A)DMM2 Reading (mA)
1.001.000.9970.0000.986
2.002.001.9920.0001.972
3.003.002.9950.0002.966
4.004.004.0000.0003.962
5.005.005.0020.0014.956
6.006.006.0000.0025.946
10.010.09.9940.0069.907
15.015.015.000.01114.876
20.020.0020.000.01519.84
25.025.0025.000.02024.821
30.030.0030.000.02529.801

 

I found the user interface fairly easy to use once I got used to it.  Setting a value is not as quick as rotating the potentiometer on my other bench power supply but it is WAY MORE PRECISE with the encoder on this unit.  Voltage is easily set to within 0.01 volts and readout is quite accurate as can be seen from the table.  A somewhat minor complaint since this is an entry level supply is that settings and readout are not displayed at the same time.  To get a setting it is necessary to push the adjustment knob which then displays the set point. 

 

Current adjustment can be done down to 1 mA, but readout and setting accuracy is not as accurate as voltage.  As can be seen from the table (note that PSU readings are in A while DMM2 is in mA), readout does not occur until the current approaches 5 mA measured with the DMM.  Readout accuracy is not given in the specifications, and this is much better performance than my other bench PSU but it was still a bit of a disappointment.  I decided to test with a smaller resistor in order to see how it performed at higher currents.  In the following table a 10 ohm resistor (measured to be 10.05 ohms) was set up as described above.

 

PSU Setting (V)PSU Reading (V)DMM1 Reading (V)PSU Reading (A)DMM2 Reading (A)
0.500.500.4990.0450.045
1.001.000.9960.0950.093
2.002.001.9900.1930.192
3.003.002.9930.2920.290
4.004.003.9970.3890.390
5.005.004.9980.4880.487
10.010.09.9860.9740.970
20.0  *** Fan turned on20.019.981.8581.853

 

The fan turned on at 20 V and roughly 2 amps.  The resistor I was using was not rated for this load and I halted the test after a few seconds.  However, as can be seen voltage setting and readout accuracy are good throughout the range, as is the PSU current reading. 

 

My main interest is performance at lower voltages and  low current so I hooked up the following circuit:

image

Basically I want to turn up the voltage slowly with the LED in series at various current control settings to see how accurate the current control setup and current readout accuracy are.  For these readings the DMM2 probes were not exactly as shown in the sketch above but were across the LED leads and not the terminals on the PSU.  Accordingly the voltage readings between the PSU and DMM1 are off a bit.  The item of interest here is the current setting and reading.

 

 

PSU CC SettingPSU Voltage ReadingDMM1 Voltage ReadingPSU Current ReadingDMM2 Current Reading
1 mA1.62 V1.62 V0 mA0.002 mA
1 mA1.65 V1.646 V0 mA0.006 mA
1 mA1.85 V1.841 V0 mA0.981 mA
1 mA1.89 V1.875 V0 mA2.241 mA
1 mA1.91 V1.889 V0 mA *** Constant Current3.016 mA
2 mA1.95 V1.910 V0 mA *** Constant Current3.109 mA
3 mA1.91 V1.889 V0 mA *** Constant Current4.825 mA
4 mA1.96 V1.922 V2 mA5.994 mA
4 mA1.98 V1.9273 mA *** Constant Current6.612 mA
5 mA1.97 V1.936 V4 mA *** Constant Current7.638 mA
6 mA1.99 V1.944 V5 mA *** Constant Current8.530 mA
7 mA2.00 V1.951 V6 mA *** Constant Current9.455 mA
8 mA2.01 V1.956 V7 mA *** Constant Current10.254 mA
9 mA2.04 V1.963 V8 mA *** Constant Current11.270 mA
10 mA2.05 V1.968 V9 mA *** Constant Current12.126 mA
15 mA2.10 V1.998 V14 mA *** Constant Current17.465 mA

 

Observe the first column which has the current control setting (mA) and the last column which has the current observed (mA) on a DMM.  Ideally, the current control setting will always be at or below the DMM current reading.  In actuality it exceeds it by up to 2.5 mA or so.  This is within the specification sheet value of 0.5% + 5mA.  Now consider the next to the last column which shows readout of current on the PSU and also whether it was in constant current mode.  The PSU does not start reading current until it reaches almost 5 mA (an error of approximately 5 mA).  The largish error exists throughout the low range.  As noted above the specifications do not specify readout accuracy for current.

 

Controlling the Unit

 

Control using the encoders is a bit different than the standard coarse / fine potentiometers and is probably best shown with a video.

You don't have permission to edit metadata of this video.
Edit media
x
image
Upload Preview
image

 

Conclusions

 

I am quite pleased with the Tenma 72-268572-2685 and recommend it for those looking for an inexpensive entry level bench DC power supply. Setup is easy and resolution is very good for an inexpensive power supply.  The unit met or exceeded the specifications in the areas I tested. The 72-268572-2685 current control does not perform as well as I would like at very low currents but it meets the specifications and the specifications are superior to the other low cost units I looked at and match some higher priced units in key areas. The power supply seems well built and was an absolute bargain at the discounted price of $68.02 at which it was purchased.  This is only about $10 more than I paid for my "eBay special" and there is no comparison in terms of performance or quality.  

 

There are higher priced models in the series for those needing higher current output, higher voltage output, and USB / RS2332 interfaces. 

 

 

 

Useful Links

Tenma 72-2685 at Newark

Tenma 72-2685 Datasheet

  • Sign in to reply

Top Comments

  • three-phase
    three-phase over 7 years ago +5
    Great blog on your new power supply, hope it gives you many years of service. I have seen encoders for control on oscilloscopes before and do like them, never seen them on a power supply like this though…
  • genebren
    genebren over 7 years ago +4
    Frank, Great write-up on your new power supply. Well done! Gene
  • shabaz
    shabaz over 7 years ago +4
    Hi Frank, Nice review : ) I've got the almost-identical model 72-2705 72-2705 , internally it looks the same, I've had mine unscrewed on my desk for the past week because I wanted to make a mod to it.…
  • fmilburn
    fmilburn over 7 years ago in reply to jw0752

    Thank you John,

     

    It is a nice power supply but the story is not as interesting as the ones you write up on your builds :-).

     

    Frank

    • Cancel
    • Vote Up +2 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • jw0752
    jw0752 over 7 years ago

    Hi Frank,

    Thanks for taking the time to give us a report on this power supply. It is always helpful to have some personal insights.

    John

    • Cancel
    • Vote Up +4 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • fmilburn
    fmilburn over 7 years ago in reply to three-phase

    Hi Donald,

     

    Thanks!  I have been using them in my projects instead of potentiometers recently and like the control they give.  It is up to the firmware somewhat and avoiding bounce...  I remember a car radio I had when they first started putting encoders in them for volume and it was awful.  I could never get the volume right :-)

     

    Frank

    • Cancel
    • Vote Up +4 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • fmilburn
    fmilburn over 7 years ago in reply to koudelad

    Hi David,

     

    First of all I have to say that I have some hearing loss so  Shabaz probably has better perspective than me :-).  At the moment my 3D printer is running and it completely drowns out the power supply, even when I draw some current from it and the fan is turning smartly.  It is  more noisy than my other power supply.  Overall quite acceptable in my opinion.

     

    Frank

    • Cancel
    • Vote Up +3 Vote Down
    • Sign in to reply
    • More
    • Cancel
  • three-phase
    three-phase over 7 years ago

    Great blog on your new power supply, hope it gives you many years of service. I have seen encoders for control on oscilloscopes before and do like them, never seen them on a power supply like this though.

     

    Be interesting to see what quality they are and how long they last.

     

    Kind regards.

    • Cancel
    • Vote Up +5 Vote Down
    • Sign in to reply
    • More
    • Cancel
>
element14 Community

element14 is the first online community specifically for engineers. Connect with your peers and get expert answers to your questions.

  • Members
  • Learn
  • Technologies
  • Challenges & Projects
  • Products
  • Store
  • About Us
  • Feedback & Support
  • FAQs
  • Terms of Use
  • Privacy Policy
  • Legal and Copyright Notices
  • Sitemap
  • Cookies

An Avnet Company © 2026 Premier Farnell Limited. All Rights Reserved.

Premier Farnell Ltd, registered in England and Wales (no 00876412), registered office: Farnell House, Forge Lane, Leeds LS12 2NE.

Follow element14

  • X
  • Facebook
  • linkedin
  • YouTube