Test & Tools
Test & Tools
Forum Worklog : Repair of 20-year old precision multimeter, HP 3458A
  • Replies 31 replies
  • Subscribers 387 subscribers
  • Views 11517 views
  • metrology
  • restoration
  • 8.5-digit
  • keysight
  • hp
  • rusty
  • repair
  • bench_multimeter
  • gear
  • multimeter
  • service
  • teardown
  • 3458a
  • agilent
tin_xdevs
tin_xdevs over 11 years ago
3458A. People who are into precision gear, need no introduction.
I wanted one since long ago, but not having business need in it, was not worth buying, as these even used on secondary market are going for good 3000-5000$ USD. But when I saw one dead unit for 750$, I could not resist and bite it. Seller photos were quite scary, with rust all over steel frame, but PCB boards looked "okay", so I hope for the best.
After some good money for shipping (beast is 15kg weight and size of average 2U rackmount server), I got it and did initial inspection:
* U100 processor MC68HC000P8 had paper sticker BAD (A5 board manufactured 1990)
* Missing button cap and lever for on power switch, guard switch and front/rear switch.
* Missing fuse binding post for amps on front panel
* Rust on steel chassis
* Mark "X" on mains power transformer
* Mark "X" on analog DC board 66501 (manufactured 1996)
* Error message RAM 1 LOW
* Missing voltage reference board (so calibration is meaningless, even if it's still intact in 1990 year DS1220)..
* Missing fuse holder for mains
* Damaged plastic rear panel
* No legs
* Dead fan
* Likely dead NVRAMs and lost cal as they are way too old.
Smell like lots of work, lots of debug folks.
Let's see what we are dealing with..
Always sad to see such instrument in mistreated condition. 3458A is current model, and you can get brand new one from Keysight, if you ready to fork 9.5K$, in base configuration.
Let's see what's inside..
Cover's removed:
View on both sides without covers. As we can see, no A9 voltage reference with Linear LTZ1000 present. New A9 reference module from Keysight runs for 700$, but hey, I don't need go that way. I had own LTZ1000(A) based references, on which I spent multiple hundreds, and on second source there is eBay, where these references popup for sale like mushrooms after rain. I snagged one of those already , should arrive somewhen next week. So we covered here..
Inguard PSU. Some PCB damage around CR12 visible.
Outguard primary PSU. All seems ok, but need cleaning..
68000 CPU, RAM, ROM and digital logic, board A5. Dated 1990, while rest of analog PCBs dated 1996.
Probably seller changed digital PCB afterwards?
Removing inguard PSU for now, adding jumpwire instead of fuse (don't have suitable cap), turning on...
It's works!.. Well, not really image Analog side completely disconnected and unpowered at this moment, just check digital and power sections first.
Nice and bright VFD, which is good, as it's hard to get replacement for those.
Now will remove Dallas NVRAMs , and solder sockets instead, so we can troubleshoot this RAM 1 LOW.
Front panel must be removed in order to remove mains transformer.
Looking good, just minor cleaning required..
Frontpanel PCBA label and front binding post terminals.
Now, we can take transformer out. I tried to open it up, but it's all covered in lacquer, and my patience run out quickly after wasting a hour trying to take it apart.
Transformer have two winding sections, primary with secondary for outguard +5V rail (goes to digital board and front panel, that's why it's working).
Second section for inguard supply is inner bobin with two secondary windings (one with center tap to generate +18/-18 and second for +5V analog).
This windings in my transformer are shorted together!
I would not trust repaired/rewinded transformer in such an instrument anyway, so going to make Agisight richer for $337 to get new transformer.
In case someone got FALSE impression that buying dead HP 3458A and fixing it would be cheap way to get yourself 8.5-digit DMM, think again.
Here are expenses on my unit already so far:
image.
Add calibration (let's say you go full tilt, standard's grade cal $2660), and you looking into healthy 4.4K$, for which you can get working and likely even calibrated 3458A from second-market, if you patient and shake trees good enough.
Also you can see I cut few corners here and there with parts from eBay, as if you go with all original parts from Keysight, it will be much more (Reference board alone is 700$, like mentioned before).
Since I wait for new transformer to arrive, how can we test inguard supply to confirm it's OK?
I have a "solution". From previous Keithley Model 2001 repair projects I have one extra power transformer, which happen to have same voltage outputs, as HP one, just at lower power rating. But since we connect only inguard supply A4 board, we don't need full power to check LDO's  operation. So let's replace blown zeners and test A4 PCBA with properly wired Keithley mains TR-280 transformer. Never know which parts could ever come handy, eh?
But first, replaced original CR11,CR12 with fresh zeners (got 50pcs from eBay, just few dollars with free shipping).
Cleaned board both sides, everything else look intact and good.
Now can connect input IEC plug socket to transformer's primary (since I need 110VAC, I used WHITE + BLACK wires).
Keithley used same type of 5-pin connector, but with different pinout. Secondary connector need rewiring as on photo above. I powered transformer separately and measured VAC voltages on secondary, without A4 board connected, to make sure all voltages correct.
If you see schematics of A4 board in 3458A Multimeter Component Level Repair Manual, you may noticed MH1,MH2,MH3 connections. These are connected directly to guard frame in 3458A's chassis, which is acting as mecca star point for GND potential. Since I testing board separately, not mounted to chassis, I needed to connect these MH* points together with copper wire. This will be our ground point for measurements as well.
Apply power to transformer, and if nothing smokes, test output DC voltage.
There are nice labels near test points, telling us voltage test points location.
All voltages are OK, also 0.325VAC 60Hz signal to read mains frequency is OK too.
Repair of A4 inguard power supply now complete.
I also got used bezel off eBay in good condition, to save some money on this project, as it's just mechanical part to keep meter in one piece.
New one on bottom side, old busted rear bezel is on top.
image image
Original unit's bezel had three out of four screw mounts destroyed, so it was not holding well.
Stay tuned, to be continued... image
  • tin_xdevs
    tin_xdevs over 11 years ago +2
    Alright, here's next update, for those who follow A5 board Replaced original rusty GPIB connector with fresh one from dead 33120A PCB, I had around. Funny traces are PCB spark-gaps for some ESD protection…
  • tin_xdevs
    Everything assembled now, and ready for initial testing.. Transformer data for comparisons: Voltage rail name Nominal spec Regulator Bad transformer in our unit New good transformer winding +18V +18VDC…
  • tin_xdevs
    tin_xdevs over 11 years ago +1
    Made an order on Digikey for fan, caps to recap A6 outguard power supply board (can call me paranoid, but I don't trust 20 year old electrolyte capacitors.). Only gotcha is that there was no axial type…
  • gadget.iom
    gadget.iom over 11 years ago
    Nice detailed article. Thanks! image
  • tin_xdevs
    Made an order on Digikey for fan, caps to recap A6 outguard power supply board (can call me paranoid, but I don't trust 20 year old electrolyte capacitors.). image
    Only gotcha is that there was no axial type C8 8300uF 35V capacitor, so I got usual one of similar size instead, will bodge something up.  image
    Also one thing catch my eye image :
    See anything suspicious?
    Hint-hint : P3 connector goes directly to fan... image
  • tin_xdevs
    Missing board A9 arrived yesterday, now we have an 8ppm/year voltage reference!
    PCBA shots:
    image image
    Vishay foil Resistors look
    image image
    And quick check if it's working correctly, with Keithley 2002 as measuring DMM and Keithley 2400 as power source.
    About +7.07421 VDC @26.5°C ambient. My 2002 is few ~30ppm's low compared to 2001 Cal.
    One good thing about used references, they likely to be nicely aged. LTZ1000A on this one is made in 1993, in case someone miss it.
    Only better with time.
    jw0752 over 11 years ago
    Hi Illya,
    I really enjoyed following your repair process. Great pictures and explanations.
  • tin_xdevs
    Alright, here's next update, for those who follow image
    A5 board
    Funny traces are PCB spark-gaps for some ESD protection on interface. Never see such on interfaces ports before, but it's more common on mains SMPS units.
    IPA+water bath for board to clean 25-year old dirt.
    image image
    A6 board
    Removing VFD supply boardie:
    image image
    It's a TDK-CD1865B DC-DC/DC-AC module
    Now since monday I will have my Digikey package with new caps, I removed old ones. They looked alright, so most of you guys would think - hey, those caps probably okay, don't waste your time?
    Well, now we see it was a good idea to remove capacitors... Some already started leaking. Time to inspect PCBA:
    Small ones did not do any damage (yet), but bigger ones...
    image image
    HP 3458A using only 2-layer PCBs, so I'll just clean it to shiny copper, add thermal-resistant tape around pins. Caps I bought are long-life +105°C, so should work a treat for next 10 years.
    Also ESR test did not show much problem with capacitors, measured by my repaired 4263B
    Verdict: caps are only getting bad with age, not good like voltage references. 10 years passed? Replace them!
    Here is timelapse of whole process: HP 3458A A5&A6 boards repair project timelapse
  • tin_xdevs
    Voltage rail name
    +18V  +18VDC  U1, LM317T    Vmeas : 7.3 VAC, 0.5 Ω  Vmeas : 20.1VAC, 1.42 Ω
    -18V  -18VDC  U2, LM337T    Vmeas : 6.9 VAC, 0.5 Ω  Vmeas : 20.1VAC 1.42 Ω
    PSS   return for +-18VDC     Vmeas : 14.2 VAC, 1.0 Ω  Vmeas : 40.1VAC 2.88 Ω
    Resistance between ±18V winding and +5V    0.5 Ω  >10 GΩ
    -18unreg     CR01-CR04 diode bridge
    +5V  +5VDC  U3, UA7805    Vmeas : 7.0 VAC, 1.0 Ω  Vmeas : 9.1 VAC
    -20V  -20VDC  CR17 zener + Pass PNP Q2
    Resistance between ±18V winding and secondary shield    0.8 Ω Shorted!  >10 GΩ
    Resistance between ±18V winding and secondary 5V shield  >10 GΩ  >10 GΩ
    And this is how transformer made:
    Assembly timelapse :
    Front panel reports REQUIRED ACAL ALL, then quickly passing TEST RAM, TEST CHECKSUM and advances to TEST HARDWARE.
    After about 5 seconds unit have one short beep and error message ISOLATOR FAILURE. This meaning digital side cannot talk to digital side of inguard section via isolation interface.
    Here’s simplified block diagram of what interfacing look like:
    If there is problem with either ASIC, it would be very hard to find replacement without buying new expensive boards. Let’s try troubleshoot issue on circuit level first, and hope for the best.
    Optical cables are simple plastic fiber, and pretty easy to check, since used receivers and trasmitter are using 660nm light, which is visible red.
    I can see red light on transmitter, and swapped cables without effect, having same error. When meter stuck on ISOLATOR FAILURE message it tries to send data, activating J102 Transmitter. If constant 660nm light directed on J101 receiver during meter power on, it resets itself in loop right after testing RAM, so likely that receiver path on A6 board is working.
    After some logic probing with my ancient Tektronix TLA714, I got idea that A5 is working okay, and problem narrow down to J102 transmitter interface
    Signal path section on schematic of A6 PCBA:
    One thing I notice before, LED light on J102 is really dim, could be also show actual problem with transmitter diode. Since it was easy, I actually swapped J102 with J303 before (which used for external triggering BNC ports), but result was same, so could be both transmitters are faulty.
    To test this I had bypass transmit isolation path by adding a shorted between anode of A6 CR101 diode to U302 pin 9 input on A3 A/D board. Guess what, it booted and let me to UI.
    image image
    image image
    It does lit up ERR with next errors:
    202,“HARDWARE FAILURE — SLAVE TEST: CONVERGENCE”
    110,“CALIBRATION REQUIRED — ACAL”
    If I try run ACAL ALL, it says 205,“HARDWARE FAILURE — CAL VALUE OUT OF RANGE: 72” which is DCV A1 value, as per ALRM document.
    That's some progress, if you ask me.
    More to come, stay tuned...
  • Robert Peter Oakes
    Robert Peter Oakes over 11 years ago in reply to tin_xdevs
    Very impressive work so far, keep it up, im rooting for you to crack this thing
    peter
    jw0752 over 11 years ago in reply to tin_xdevs
    Hi Illya,
    This is better than a good mystery novel. I am eagerly awaiting the next chapter.
  • tin_xdevs
    Fixed optical isolator issue, it was dead receiver on inguard A/D converter board. Swapped it with receiver from EXT trigger section (I don't need that now, will buy replacement part later) and bingo, it boots with optical isolation. image
    And guess what, yesterday was first time I ran out of hooks for LA... used total 21 inputs (out of 136 possible on my TLA7AA4 image)
    Decided to upgrade firmware for latest version... Only hardcore, only oldschool, using by bodged lamp to generate UV for UVEPROM erase, so I can flash chips.
    Next steps
    A1 DC board debug.
    Current diagnosis:
    DCV (any range) reading as near zero, but should be bouncing around on 10VDC and lower as it's >10GOhm impedance. So something somewhere cause too low impedance/short.
    Ohms reading near zero for 10K and higher ranges, but OVLD for lower.
    Current reading something microamps.
  • tin_xdevs
    Forgot jumper near voltage reference, image
    Now meter behaving much much better and can get first reading out of it!
    Measuring +10.0006 VDC from SMU:
    image image
    image image
    Ohms mode
    image image
    image image
    DCI ranges 10nA, 100nA, 1uA, 10uA, 100uA showin zero, no matter what current supplied into meter. So those ranges are broken. That cause self-test fail on value 95 as well.
    DCI ranges 1mA and up are fine.
    So happy now.
    Video with Self-test, ACAL attempt and some tests.