One of the easiest way to exercise the analog frontend of the Red Pitaya STEMlab 125-14 PRO Gen 2 is by feeding it a known signal and see how it is perceived. For this,t he oscilloscope app is the best candidate.
After starting the app, I plugged the probes to the 1kHz 5V reference signal for probe compensation on the MP720012 that I have on my bench. I was expecting to see this:

Instead, I got this:

The manual says this:
Red Pitaya boards are factory-calibrated before shipping and should provide accurate measurements immediately after purchase.
But clearly, this is not what I’m seeing here with range being clipped, so I decided to go through the whole calibration process:

Values were a bit off, but after that, nothing changed on the oscilloscope!
After a bit of search, I found a clue in the forums: https://forum.redpitaya.com/viewtopic.php?t=1128
And with the proper “jumper” search term, one can finally find the information in the documentation, deep down in the developer’s guide hardware section: https://redpitaya.readthedocs.io/en/latest/developerGuide/hardware/GEN2/hw_specs/hw_specs.html#jumper-settings
So, by default, the input stage crops out anything outside -1/+1V and one has to move a pair of jumpers to the HV position to get -20/+20V. But the application guide for the oscilloscope does not mention this in any clear way, nor does it mention the absolute voltage value which is found in yet another page: https://redpitaya.readthedocs.io/en/latest/developerGuide/hardware/GEN2/125-14_Gen2_Pro/top.html#technical-specifications
This means that I could have “fried” the board with giving it what I thought is a perfectly innocuous test signal. A warning on the oscilloscope documentation page is more than required.
The jumpers have an orientation, that’s a first for me and quite a surprise. And clearly, what’s mentioned in the documentation about trace impact is right:
Incorrect jumper placement can cause the front part of acquired square wave signals to be overshot or undercut.
I did not find this early one and spent quite a while turning the probe adjustment knob, to no avail.
But really, it’s totally inconvenient to have those jumpers, they are VERY tiny, difficult to take out, and prone to land on all the components on the PCB. A regular jumper would not be a danger, but with the metallic part flush at the bottom, there is a risk of making contact with something on the board.
And we can’t really turn off the board while changing their position because the calibration process would then start over after reboot.
Also, it is mentioned that both jumpers should be on the same side of the connector. Well, it would be much easier if they were “glued” together, thus preventing the user from even being able to create a forbidden configuration with undocumented consequences.
For ease of manipulation, they also should have a “long tail” or replaced altogether by a DIP style two-way switch. Sure, it may take more board room and require a toothpick to manipulate, but at least, it won’t jump out of my desk and require me to spend a good 5 minutes to finally find it between the floor and the desk leg.
This blog is part of 7-part series for my Red Pitaya STEMlab 125-14 PRO Gen 2 Starter Kit roadtest:
Blog 1 – Unboxing the Red Pitaya STEMlab 125-14 PRO Gen 2 Starter Kit
Blog 2 – First contact with the Red Pitaya STEMlab 125-14 PRO Gen 2
Blog 3 – Oscilloscope application in the Red Pitaya STEMlab 125-14 PRO Gen 2
Blog 4 – Logic analyzer application in the Red Pitaya STEMlab 125-14 PRO Gen 2
Blog 5 – Software defined radio with the Red Pitaya STEMlab 125-14 PRO Gen 2
Blog 6 – Developing for the Red Pitaya STEMlab 125-14 PRO Gen 2
Red Pitaya STEMlab 125-14 PRO Gen 2 roadtest report