Test & Tools
Test & Tools
Documents Programmable Electronic Load - Analyse the Summing Node Zero Point
  • Date Created: 1 Dec 2017 4:08 PM Date Created
  • Last Updated Last Updated: 15 May 2020 3:38 PM
  • Views 12969 views
  • Likes 8 likes
  • Comments 107 comments
This blog documents investigates the feedback node of the electronic load that Robert Peter Oakes, jc2048 and Jan Cumps are designing.
It's an important spot in the load's design. It measures the set point and the feedback from the output.
When the output is driven to 0, it should be on a potential as close as possible to 0 V.
On the first prototype it's -0.2 V. Not so much off, but the negative value  influences our ADC measurements.
This document checks how we can get this node to 0 V.
Because this document is evolving, some comments below may be out of sync with the content. That's because the content is adapted based on the conversation.
The measurements taken here are based on the original design, without R32 in place and U3B + tied to ground.
The current sense side of R7 is connected to ground, and a variable negative voltage from 0 V down is applied to the current sense side of R8 to simulate current being sensed.
The circuit isn't complex. The set point is driven by a DAC. It's set to 0 for this test.
The second input to this node is OpAmp 3C. It has both inputs tied to ground so should theoretically have 0 V at the output.
On my board I measure a potential of -0.212V at the left side of R33.
I hope to get this closer to 0 V to ease the ADC a bit - its performance degrades with negative voltage at its inputs.
Like the other blogs for the electronic load, this is a working document that will be updated with findings from anyone who wants to chime in.
Behaviour at 0V
buzy image
  • metrology
  • differentiator
  • boosterpack
  • opamp
  • msp432
  • launchpad
  • analog
  • laboratory
  • instrument
  • jc2048
    jc2048 over 8 years ago in reply to Robert Peter Oakes +4
    As you say, changing the op-amp is one possibility. There are bipolar op-amps with much lower bias currents, but an alternative nowadays is a precision CMOS op-amp (you'd have to check whether other characteristics…
  • Robert Peter Oakes
    Robert Peter Oakes over 8 years ago +3
    To hopefully simplify things a little We have this, Upper op amp is simply to provide an inversion of the measured value back tot he ADC, hence the gain of -1 Lower right op amp measures the volts across…
  • jc2048
    jc2048 over 8 years ago in reply to Jan Cumps +3
    Love the advertisement for "John's excellent probes". It's like one those things from the old days of American TV where the presenter would suddenly turn, look very earnestly at the camera, and start reading…
    I just noticed that you have a 680K going from the UC3 + to ground, this should not be populated and will throw things off in the real circuit when the common mode volts start being applied etc.
    Jan Cumps over 8 years ago in reply to Robert Peter Oakes
    I've just put it in for the exercise in this thread. I can easily remove it again (happy that I didn't go for crazy 0402 parts on this PCB).
    this isn’t populated. it’s there for experimental reasons but not part of our BOM.
