Peter, Jon and Jan are building a programmable electronic load. This document is the common design sheet.
It's obviously work in progress. And fun.
We're building a programmable DC load. The focus is on making a real world working instrument.
Hardware and firmware are open source.
  • 20 V, 6 A
  • SCPI controlled
To keep this document readable, this main post gives an overview of the design.
All detailed explanations are broken out in separate posts.
Hardware
The load design is modular.
Two out of box modules:
  • MSP432 LaunchPad
  • LCD Shield (optional)
3 new designs
  • DAC-ADC module
  • Analog control module
  • Power module
DAC-ADC BoosterPack
This module is a generic, isolated DAC / ADC boosterpack. It's the interface between the analog controller and the microcontroller.
It can also be used outside this project, for your other analog - digital conversion needs.
The board design is broken out in a separate post: Programmable Electronic Load - DAC / ADC BoosterPack.
Analog Controller and Driver Board
This is the core of the analog constant current block. Here lives the circuit that will translate desired constant current into the driver signal for the power MOSFET.
It integrates the feedback coming from that power board. This is a closed analog control loop.
There are also components to provide the on-off functionality and buffers/amplifiers for the analog data-points that fly back to the ADC module.
The board, schematics and BOM are documented in Programmable Electronic Load - Analog Controller and Driver Board.
Power Module - the MOSFET Board with Fan
For details  of the design of the power stage, check Programmable Electronic Load - Power Stage.
BoosterPack Header Use and Pin Allocation
This info is broken out in a separate post: Programmable Electronic Load - BoosterPack Header Use.
Firmware
Prerequisites:
  • CCS 8 with MSP432 compiler TI v 16.12.0.STS or >
  • TI-RTOS for MSP43X, any SimpleLink version
  • MSP432 LaunchPad (not the black pre-prod because it's out of support)
The firmware is for the MSP432 microcontroller. But the harware can be driven from any controller, processor, PSoC or computer that can deliver i2c and power between 3V3 and 5V.
That includes Arduino, Raspberry Pi, BeagleBone, Warp7, IOT20xx, the ST Neo family, ... As long as they match that voltage range and have I2C master, it 'll work.
The ADC/DAC board takes care that the logic side is galvanicaly isolated from the power side (that is, if you leave the P1 and P2 jumpers open).
The only requirement is that you have a positive supply between 3V3 and 5V, a ground, and I2C signals (the board has I2C pull-ups that connect to the positive supply that you provide. If you don't want that (maybe your dev board already has 'm), leave R1 and R2 off.
ADC, DAC and Enable are all I2C driven.
GITHUB location: https://github.com/jancumps/msp432/tree/master/MSP432_SCPI_ElectronicLoad
download latest source files as a zip : https://github.com/jancumps/msp432/archive/master.zip
doxygen generated API documentation: https://jancumps.github.io/msp432
GIT artifacts don't contain CCS project files, only sources - get a zip with CCS project from the attachments at the end of this document.
SCPI interface and Error Handling
The SCPI commands are documented in this breakout post: SCPI interface.
UART Configuration
Together with SCPI, UART is the programming interface of this instrument.
The approach, and how to compile for USB or TTL, is documented in this breakout: UART.
RTOS and Posix Configuration
The TI-RTOS documentation is in breakout post RTOS Config.
LCD Display
Check here for a detailed explanation: LCD Display
The firmware supports SHARP LCD display,
ADC
I've broken out the ADC description to a separate post, because I'm working on improvements:
Programmable Electronic Load - ADC Firmware
DAC
DAC output is set on command, using RTOS MailBox and messaging.
The payload for a DAC message contains the DAC channel that needs to be set, and the value.
typedef struct MsgDAC {
    uint8_t module;
    uint16_t value;
} MsgDAC;
The DAC task is started by RTOS. It inialises the DAC settings, then waits until it receives a message.
void *threadDAC(void *arg0) {
    MsgDAC d_msg;
    d_daci2cTransaction.writeBuf = d_dactxBuffer;
    d_daci2cTransaction.readBuf = d_dacrxBuffer;
    d_daci2cTransaction.slaveAddress = DAC_I2C_ADDR;
    d_daci2cTransaction.writeCount = 3;
    d_daci2cTransaction.readCount = 0;
    mqd_t mq;
    struct mq_attr attr;
    attr.mq_flags = 0;
    attr.mq_maxmsg = 1;
    attr.mq_msgsize = MSGDAC_SIZE;
    attr.mq_curmsgs = 0;
    mq = mq_open(QUEUE_NAME_DAC, O_CREAT | O_RDONLY, 0644, &attr);
    while (1) {
        ssize_t bytes_read;
        bytes_read = mq_receive(mq, (char *)&d_msg, MSGDAC_SIZE, NULL);
        /* wait for mailbox to be posted by writer() */
        if (bytes_read) {
The waiting doesn't take processor time. RTOS takes care that it will only get a slice of clock cycles when there is a message.
Two things in the RTOS configuration make this happen:
The mailbox with room for exactly one message, and a receive event.
Because there is usually no message in the mailbox, the execution of the DAC logic becomes inactive at this line:
When we send a DAC message somewhere else in the code (e.g.: when the user requests a new setting of the electronic load), RTOS reactivates the process and hands over the payload message.
The DAC task sets the output of the requested channel and returns to the point where it waits for a new message.
   while (1) {
        ssize_t bytes_read;
        if (bytes_read) {
            d_dactxBuffer[0] = getAddressFromModule(d_msg.module); // set value direct
            d_dactxBuffer[1] = d_msg.value >> 8; // MSB
            d_dactxBuffer[2] = d_msg.value; // LSB
            if (! I2C_transfer(i2c_implGetHandle(), &d_daci2cTransaction)) {
//                System_printf("I2C Bus fault\n");
//                System_flush();
You select the DAC channel by setting bit 2 and 1 in the control byte (tx_buffer[0]). There's a helper function that gives a correct control record.
// address 7 - 6: 0, load mode 5 - 4: 01 direct from i2c, 3: reserved 0, 2 - 1: channel select, 0: pwr down 0
#define DAC857X_CFG_H0 0b00010000
#define DAC857X_CFG_H1 0b00010010
#define DAC857X_CFG_H2 0b00010100
#define DAC857X_CFG_H3 0b00010110
static const uint8_t array_DAC857X_CFG_H[4] = {DAC857X_CFG_H0, DAC857X_CFG_H1, DAC857X_CFG_H2, DAC857X_CFG_H3};
 * get the hex address for the requested DAC module
uint8_t getAddressFromModule(uint8_t module) {
    return array_DAC857X_CFG_H[module];
Bits 7 - 3 and 0 are all fixed.
Input Enable
The functionality to activate and deactivate the load is documented here: Programmable Electronic Load - Input Enable Functionality .
Control Strategies
The firmware uses strategies to handle the different types of load operation (e.g. constant current, constant, voltage, ...).
A strategy is a set of functions that you can plug in, and that together run that particular operation type (this is a c version of the Gang of Four's Strategy Design Pattern).
The goal is to avoid that the firmware is riddled with if or switch statements whenever different behaviour is needed depending on the instrument's mode.
Each operation strategy will have the same set of functions that run the instrument in that mode.
The price to pay is not expensive - speed and memory burden is low. It is a little more complex to understand than the c++ version. Once you step trough it with a debugger, things become clear.
The code also tries to hide that we're using strategies
Currently starting to implement constant current strategy. The more difficult operation types can be added later, one by one, with not too much impact on the existing code across the firmware.
. Only the eload API (see below) knows about it. All code goes via that eload API.
API for the stategies is minimal now:
    controlFunction;
    getMode;
    getChar;
eload API
The eload API offers an abstraction layer for the strategies, so that the rest of the firmware doesn't have to know about it. It simplifies switching operation mode and driving the instrument.
Example: The eload API offers a simple function eloadGetMode() to check what the instrument's current mode is.
eload_mode eloadGetMode() {
    return getControlStrategy()->getMode();
In the background, it uses the strategy mechanism to fetch the mode from the currently active strategy and call the implementation function.
The eload API also has the common eload functions that don't need a strategy because they are always valid, regardless of mode.
Calibration and Configuration Data
The functionality to store calibration and configuration data in Flash is described in a separate article: Programmable Electronic Load - Calibration Data in Flash.
The calibration and configuration procedures are documented in Programmable Electronic Load - Calibration Steps
Changes To Standard MSP_EXP432P401R.c
The standard file generated by the TI-RTOS application wizard has defaults for peripherals.
The following has changed: none - I removed this section because with the external ADC/DAC board and the SimpleLink RTOS version, everything is kept standard.
Remote Software
Teminal
Serial communication to the USB UART port.
9600/8/1/None
SCPI is a non-echoing protocol. To make the characters you type show in the Terminal, set Local echo to Force on.
If you also set Local line editing to Force on, you have the opportunity to correct characters in the current line before they are sent to the serial port.
Test:
*IDN?;SYST:ERR:COUN?
Should return
THEBREADBOARD,ELECTRONICLOAD,0,01.00;0
Windows GUI
There are two image. A .NET made by Peter, a Java version made by Jan.
The .net GUI supports reading the 4 ADCs, setting one of the DACs abd shoot any SCPI command.
There's a window for SCPI output and a status window that shows SCPI errors.
Source and binary available from https://github.com/thebreadboard/SCPI_DC_LOAD_WIN
The Java GUI is made as a POC that cross-platform instrument GUIs are possible.
It allows for switching the load on and off, set constant current and replicate the LCD values.
Additionally, you can requesst error status and run arbitrary SCPI commands.
Programmable Electronic Load - Java GUI Part 1: Basic Functionality
LabVIEW
The instrument comes with a LabVIEW library: Programmable Electronic Load - Write a LabVIEW Library part 1: Initialise Block
Related blog
Programmable Electronic Load - Temperature Protection
Programmable Electronic Load - SCPI interface and Error Handling
Programmable Electronic Load - UART
Programmable Electronic Load - RTOS Config
Programmable Electronic Load - LCD Display
Programmable Electronic Load - Current Sense Circuit
Programmable Electronic Load - ToDo and Done
Robert Peter Oakes articles:
Electronic DC Load - Design and Build to test PSU Project
youtube: Electronic DC Load Design and Testing
youtube: Full Build Analogue DC Load
youtube: Electronic DC Load - Performance Improvements
Raspberry PI 2, Fun with I2C DACs and ADC's
jc2048 articles:
Programmable Electronic Load: Dynamic Behaviour: Part 1 Overview
Programmable Electronic Load: Dynamic Behaviour: Part 2 The Servo Loop
Programmable Electronic Load: Dynamic Behaviour: Part 3 Effect of Output Inductance
Programmable Electronic Load: Dynamic Behaviour: Part 4 Effect of Output Voltage Change
Programmable Electronic Load: Dynamic Behaviour: Part 5 Stability
Programmable Load Build
Jan Cumps articles:
Programmable Electronic Load - ADC and DAC BoosterPack test
Programmable Electronic Load - Analyse the Summing Node Zero Point
Programmable Electronic Load - Measurements Part 1: Control Circuit
Java GUI
Programmable Electronic Load - Java GUI Part 2: Support for Current Mode, Input On/Off, Error Log
LabView
Programmable Electronic Load - Write a LabVIEW Library part 2: Read Output Block
Programmable Electronic Load - Write a LabVIEW Library part 3: Close Block
Programmable Electronic Load - Write a LabVIEW Library part 4: Function Set Block
Programmable Electronic Load - Write a LabVIEW Library part 5: Input Control Block
Programmable Electronic Load - Write a LabVIEW Library part 6: Raw DAC Block
Programmable Electronic Load - Write a LabVIEW Library part 7: Raw ADC Block
Programmable Electronic Load - Automating a DC Switcher Efficiency Test with LabVIEW
Programmable Electronic Load - LabVIEW Test Automation: Characterise the Instrument
Programmable Electronic Load - LabVIEW Test Automation: Characterise the Instrument Pt 2: Oscilloscope Measurements
Example LabVIEW Scenario: test a design with a Rigol PSU and a Keithley DMM
TI-RTOS: Switching to the SimpleLink Distribution
Switch from TI-RTOS to SimpleLink POSIX: Threads and Semaphores
Switch from TI-RTOS to SimpleLink POSIX: EEPROM API
Switch from TI-RTOS to SimpleLink POSIX: From MailBox to Message Queue
Switch from TI-RTOS to SimpleLink POSIX: LCD Display Driver
Switch from TI-RTOS to SimpleLink POSIX: Sleep when Idle
pcb_proto.zip
eload_v1_20170227.zip
eload_1_2a.zip
https://community.element14.com/cfs-file/__key/communityserver-wikis-components-files/00-00-00-00-21/bom.ods
MSP432_SCPI_ElectronicLoad_20180102.zip
eload_power_20180104.zip
eload.zip
eload_examples.zip
MSP432_SCPI_ElectronicLoad_20180514.zip
measureefficiency.zip
measureefficiency_5V_10V_600mA_RAW.zip
MSP432_SCPI_ElectronicLoad_v_00_0_0_1_20180914.zip
MSP432_SCPI_ElectronicLoad_20190706.zip
  • load_cell
  • metrology
  • msp432
  • ti-rtos
  • launchpad
  • laboratory
  • instrument
    Jan Cumps over 6 years ago +4
    Revisiting the eLoad design - there's definitely room for approval. The very low currency (minus 3 mA) is awful. And the calibration is off. I've set it to 0.5 A- and it thinks it's drawing that current…
    Jan Cumps over 9 years ago in reply to Jan Cumps +3
    All components are now in and soldered. Time to test. I'll first check out the power rails in isolation. If they are ok. I'll mount LaunchPad, LCD and and DAC board and do measurements. Hang on
    Jan Cumps over 8 years ago +3
    good news: I finally got persistent saving of calibration/setup sorted out. At first I had issues getting the in-ROM Flash api recognised by the compiler when using it with RTOS. That turned out to be…
    jc2048 Robert Peter Oakes, I have time to upgrade the firmware to the latest RTOS (the dreaded SimpleLink version image ).
    Is that ok?
    It would require you to run CCS 7.1.
    I've ported to the new TI-RTOS and have an issue.
    UART write only writes one character and then hangs.
    I first thought it was my code. So I tried to load one of the RTOS examples and try that.
    I have the same issue when using the TI default unchanged uartecho example.
    I opened a support ticket with TI: https://e2e.ti.com/support/microcontrollers/msp430/f/166/t/615037
    I've installed CCS7. I've got a MSP432 Launchpad and a Sharp LCD, so when you're ready with some new code I can try it out.
    The issue may be related because I have an early pre-production version of the LaunchPad.
    It would help if you could try to replicate these steps with your newer one:
    In CCS, select View -> Resource Explorer
    Install latest SimpleLink for MSP, if not done yet
    Then, from the tree view in that same Resource Explorer, load the following project into CCS:
    This will give you a new project called uartecho_MSP_EXP432xxxxx_tirtos_ccs
    Select it, then with the LaunchPad plugged in, press the debug button.
    When the debugger is loaded, open a terminal (PuTTY or whatever you like, 115200, 8 1 N , the com port is the one that represents itself as Application / User UART )
    Then press the resume button in CCS.
    If the terminal shows "Echoing characters:" the issue is with my hardware.
    If it shows "E" , then it may be an issue with the latest RTOS.
  • jc2048
    jc2048 over 9 years ago in reply to Jan Cumps
    I've got the black 'Rev 1.0' board. It looks the same as the one you show in your photographs.The processor has 'XMS432 P401R REV B' on it.
    It's what Farnell in the UK still sell, if you buy one from them:
    http://uk.farnell.com/texas-instruments/msp-exp432p401r/dev-board-msp432-performance-launchpad/dp/2473128
    I'll try it anyway this evening, but I suspect it will behave exactly as yours does.
    Jan Cumps over 9 years ago in reply to jc2048
    I got confirmation on the TI ticket that it's related to the version of our controller.
    I've archived my upgraded project. Reverting back to TI-RTOS for MSP43x 2.20.0.06
    If you want to try, there's a ZIP with the latest version attached to the main post (the 4th attachment).
    It is possible that it hangs on the I2C ADC task if you don't have the PCB with DAC and ADC mounted. In that case comment out the I2C code in the adc_impl.c file, or remove the ADC and DAC tasks from the RTOS configurator..
    Peter has made a GUI to test the board. It also works with LabVIEW or a terminal program using the SCPI commands from the main post.
    it  may be solved by installing TI-RTOS for MSP430 via View -> Resource Explorer.
    You install it via the download button, check version via the i button:
    Can you right click on the project, the select properties,
    General -> Products.
    Then check if these are selected on your installation:
    It may be related to what rtos you have installed ...
    The greyed out entries is normal. They are supposed to be excluded from build.
  • jc2048
    I'm trying to compile your code from the zip file with CCS7
    MSP432_SCPI_ElectronicLoad_20170808.zip
    I get these errors
    and the entries in Project Explorer are greyed and have a strikethrough
    Any idea what I'm doing wrong?
    Firmware now supports setting Constant Current via SCPI.
    It's still RAW DAC value. Once we know the relation between DAC output and the current that it sets in the power part, this can turn into a command that takes floating point Current .
    Official command structure:
    [SOURce:]CURRent[:LEVel][:IMMediate] <VAL>
    Practical, can be shorthanded to:
    CURR <VAL>
    Example:
    CURR 12345
    Sets the DAC A output (I'm assuming that we use the first DAC channel to drive the power part) to  REFVOLTAGE * ( <VAL> / 65535).
    As always: I can't be trusted with figures. Need to test first.
    Zip file in the main post and GITHUB have latest version.
    The DAC IC should arrive tomorrow.
    On the other news: I've attached a fresh CCS project to the main posts with latest changes:
    • 4 channel DAC support (untested for channels B, C and D - need the IC for that)
    • plug-able operation mode (see Control Strategies in the main post) - should allow us to implement different modes (constant current, constant voltage, ...) with little impact to existing code.
    I've just tested the 1_2 version of the ADC/DAC BoosterPack.
    The 4 ADC channels and VREF work fine.
    I can't test the DAC yet because I still don't have one.
    edit: ordered one DAC8574IPW
  • jc2048
    End of February, so about five months ago. It's possible that Farnell have changed to the red board since then, but the sales page still shows a picture of a black one.
    There's a small sticker on the end of the box that has "Week23/2015" on it, so it looks like they have a large batch from two years ago that they're still selling. It doesn't matter much to me if I can just use the old versions of the development software, but it's very confusing if you're new to TI, aren't up to speed with the IDE, and no-one has told you that that's what you need to do.
    I'll have another go a getting your code to compile - last night it was coming up with errors about the RTOS stuff.
    You can install the version of RTOS that we're using in this project from Resource Explorer:
    If that's installed, you can import the firmware via Project -> Import CCS Project -> Select Archive File -> Browse
    It should image compile and run on the LaunchPad.
    The support stops. As long as you have a compatible compiler (the ones with LTS behind the version number are long term support) and rtos, it'll keep on working.
    What surprises me is that the black boards are still shipped to customers. Did you get yours recently?
  • jc2048
    This page has information relating to the REVB preproduction silicon.
    http://processors.wiki.ti.com/index.php/XMS432_Support
    "...if you would prefer to use pre-production XMS432P401R Revision B silicon, you may continue to do so until June 2017."
    If I'm reading that wiki page right, I've bought a 'development board' that I can't do development with.
  • jc2048
    I don't have a SYS/BIOS option there in that box. Maybe because I've never downloaded/installed a SYS/BIOS by itself. I rely on the one embedded in the TI-RTOS release.
    The project uses this one by default on my project: 6.46.0.23.
    Can you first try to unselect the SYS/BIOS,
    Then right-click on the file MSP432_SCPI_ElectronicLoad.cfg and select Open witk XGCONF
    You will get a new screen left bottom, under the three, with "available products"
    That lists he SYS/BIOS that the RTOS version is using by default.
    If that's the same as on this image, try to do a full Rebuild of the project....
  • jc2048
    This is what I see. It looks similar to you.
    The errors are Make complaining because it doesn't have a rule for compiling BIOS stuff, whereas you think you're just linking the object files.
    If I enable a BIOS package (below), it compiles without errors.
    However, that's probably the wrong BIOS version. Which one are you using?
    .... and on e2e they say:
    - Right Click on the project and select Clean Project
    then
    - Right Click on the project and select Build Project.
    (can be done in one step if you select Rebuild Project)
  • jc2048
    I get these errors
    CURR <VAL>
    Example:
    CURR 12345
  • jc2048